主题:【求助】看看我变色酸法做的甲醛标线有什么问题?

浏览0 回复6 电梯直达
hxueren
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
变色酸0.1%,Tube No. Volume, mL
NaHSO3(1%)  Solution B(10ug/ml甲醛的1%的NaHSO3)
Std1 4.0ml              0.00ml
Std2 3.9ml              0.10ml
Std3 3.7ml              0.30ml
Std4 3.5ml              0.50ml
Std5 3.3ml              0.70ml
Std6 3.0ml              1.00ml
然后每个加入6.00ml浓硫酸摇匀,2小时后在紫外-可见分光光度计,580nm处以水为参比测试,结果如下;

Report time                    2008-12-5 15:28:11                                                             
Method                        E:\formaldehyde in furniture\method\ASTM                                       
                              D6007-2.MCN                                                                     
Batch name                    E:\formaldehyde in furniture\data\20081205.BCN                                 
Application                    Concentration 3.00(339)                                                         
Operator                                                                                                       

Instrument Settings
Instrument                    Cary 300                                                                       
Instrument version no.        9.00                                                                           
Wavelength (nm)                580.00                                                                         
Ordinate Mode                  Abs                                                                             
SBW (nm)                      1.5                                                                             
Ave Time (sec)                0.100                                                                           
Beam mode                      Double auto select                                                             
Beam interchange              Normal                                                                         
Replicates                    1                                                                               
Standard/Sample averaging      OFF                                                                             
Weight and volume corrections  OFF                                                                             
Fit type                      Linear                                                                         
Min R2                        0.95000                                                                         
Concentration units            ug                                                                           

Comments:

Zero Report

    Read            Abs            nm       
________________________________________________
    Zero            (-0.0173)          580.00

Calibration
Collection time                2008-12-5 15:29:58                                                             

      Standard      Concentration  F  Readings(ABS)
                          ug                 
________________________________________________
Std 1                      0.0000      0.0134
Std 2                      1.0000      0.0754
Std 3                      3.0000      0.1921
Std 4                      5.0000      0.3063
Std 5                      7.0000      0.4204
Std 6                      10.0000      0.5967

Calibration eqn              Abs=0.05805*Conc+0.01586                                                     
Correlation Coefficient        0.99993                                                                       
Calibration time              2008-12-5 15:35:47                                                             

Analysis
Collection time                2008-12-5 15:35:47                                                             

      Sample        Concentration  F  Readings
                        mg/L                 
________________________________________________
MB-1                      -0.0041      0.0156
MB-2                      -0.0229      0.0145
MB-3                      -0.0229      0.0145 
MB-4                      -0.0190      0.0148
MB-5                      -0.0414      0.0135
MB-6                      -0.0265      0.0143
MB-7                      -0.0286      0.0142
MB-8                      -0.0267      0.0143
MB-9                        0.0150      0.0167
MB-10                      -0.0244      0.0144
MB-11                      -0.0134      0.0151
MB-12                      -0.0360      0.0138
MB-13                      -0.0252      0.0144
MB-14                      -0.0257      0.0144
MB-15                      -0.0088      0.0153
MB-16                      -0.0159      0.0149
MB-17                      -0.0143      0.0150
MB-18                      -0.0179      0.0148
MB-19                      0.0198      0.0170
MB-20                      -0.0252      0.0144
MB-21                      -0.0277      0.0142
MB-22                      -0.0402      0.0135
Std 2                        1.0273      0.0755
Std 2                        1.0268      0.0755
Std 2                        1.0303      0.0757
Std 2                        1.0290      0.0756
Std 2                        1.0284      0.0756
Std 3                        3.0373      0.1922
Std 3                        3.0223      0.1913
Std 3                        3.0250      0.1915
Std 3                        3.0220      0.1913
Std 3                        3.0203      0.1912
Std 4                        4.9840      0.3052
Std 4                        4.9859      0.3053
Std 4                        4.9881      0.3054
Std 4                        4.9838      0.3052
Std 4                        4.9864      0.3053
Std 5                        6.9923      0.4217
Std 5                        6.9751      0.4207
Std 5                        6.9704      0.4205
Std 5                        6.9635      0.4201
Std 5                        6.9640      0.4201
Std 6                        9.9910      0.5958
Std 6                        9.9862      0.5955
Std 6                        9.9862      0.5955
Std 6                        9.9311      0.5923
Std 6                        9.9328      0.5924

这些点又重新测了,怎么误差那么大啊,从吸光度看都差不多,但是从等式算出来就不一样了,差别还比较大的?尤其是空白吸光度0.0145 ,算出来却是-0.0229,而Std 1  0.0000      0.0134; 0.0145比0.0134还大怎么是负值?
另外根据这些能不能算检出限?怎么算?我算的怎么比空白还小很多啊?     
为您推荐
您可能想找: 气相色谱仪(GC) 询底价
专属顾问快速对接
立即提交
可能感兴趣
yangjianqin
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
从吸光度看都差不多,但是从等式算出来就不一样了,差别还比较大的。
看不明白???????
你能确保在样品配制过程中是否准确?是否超仪器检测的线性范围?
赞贴
0
收藏
0
拍砖
0
2012/3/20 17:26:54 Last edit by hhx050
做牛做马
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
是不是你的仪器没有预热好啊,要不就是你每次测试的时候没有洗干净比色皿,
☆真空★
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
请问能把你适用的测试标准发个我一份吗?我也正想用变色酸测甲醛可是不知道用什么标准,发到我邮箱好吗谢谢了!邮箱地址:moreboshi@163.com
hxueren
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 kongxy1983 发表:
请问能把你适用的测试标准发个我一份吗?我也正想用变色酸测甲醛可是不知道用什么标准,发到我邮箱好吗谢谢了!邮箱地址:moreboshi@163.com

标准就是ASTM D 6007-2002
lz20052006
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
我也在用这个标准做哟!
有时间交流一下,
jinanhongwei
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
astmd6007
兄弟,给你标准
附件:
该帖子作者被版主 pjs1232积分, 2经验,加分理由:谢谢分享,欢迎发帖