We had similar problems on the 2487 before.
Here is the solution we got from Waters technical service.
Open the cover -> locate the battery inside in the detector (deuterium lamp.)
( the stored lamp data is invalid)
->After changing the battery, the lamp' data need to be reset, ( you could do it manually by using the input/output diagnostics or use the millennium to perform the diagnostics precess. ->
need to re-calibrate again. or power on/off again and press "enter" let the lighting warm up 5 mins.
Hope it helps.
soho1107回复于2009/04/21
时间久了没开机,是否是主板电池没电了,参数都丢了吧.
sulian2008回复于2009/04/21
同意!给主板换个新电池试试就知道。
liyunk回复于2009/04/22
那个是由于电池没电了,市场上修表的地方买个2032纽扣电池换上,手动校正一下就ok了,花不了几块钱,