原文由 hfang2005(hfang2005) 发表:
quote]原文由 消咳喘(ROGERSW) 发表:
我们的ECD 置于"OFF"状态时,使用FID,仪器面板显示尾吹气流为零。
数年前用过GC2010 FID+ECD,现在仪器转到外地工厂了,但是英文版说明书的COPY有一份.如果有你所说的情况,那就是故障,可以让他们修理.因为气路的开和关与ECD的"ON"或"OFF"不应联动,英文版说明书中写的清清楚楚的如下:
DETECTOR CONTROLLER
Selection: On/Off, Default: On
When a detector is set to “On”, its current and voltage are controlled.
When a detector is selected in a line and set to “On” here, the detector is used.
However,
even if a detector is set to “Off” here, the gas flow and the temperature rises and reaches the
setup value if the detector is configured in a line.不知道楼主注意没有,如果是system off时,尾吹还有没有气流?即使气路的开和关与ECD的"ON"或"OFF"不联动。