几种常用NMR实验的基本参数:
General information
SENSITIVITY
Higher field gives better S/N, so 600>500 > >400>300 MHz
Probes optimized for a particular nucleus give the best S/N for that nucleus. The TXI-500 is an indirect detection probe and is thus most sensitive to proton. The most commonly used probes are QNP-500 and QNP-400. The QNP(Quattro Nuclear Probe) has 4 channels for running 1H, 13C, 19F and 31P. Other probes are also available.
RESOLUTION and DISPERSION
Higher field spreads out the signal better, makes coupling patterns simpler, separates different signals better. Again, 600>500 > 400>300 MHz.
Linewidths in Hz are similar whatever the field, so the linewidths in ppm are smaller at the higher field strengths, unless a frequency-dependent line broadening mechanism interferes.
Proton-detected experiments
Availability
H1 experiments are available on all instruments.
1D Proton experiment
Default Parameters
~2 second acquisition time
~1.5 second relaxation delay
Spectral width -2ppm to 18ppm
16 scans
Total time: 1 minute
Variations on the 1D experiment include
1D NOE or selective NOE (difference or gradient)
1D COSY or Selective COSY (gradient)
T1 determination
Recommended Concentration
> 0.1mM
COSY - Proton-proton correlation experiment
Gives information about pairs of protons that are J-coupled. This usually indicates that the protons are on adjacent carbons, e.g., 3-bonds away (though protons further apart may in some cases be J-coupled).
Default Parameters
1.5 second relaxation delay
512 complex points in t2 and 192 increments in t1
4 scans per increment (no PFG) / 1 scan per increment (PFG)
Total time: 20 minutes (no PFG) / 5 minutes (PFG)
Recommended Concentration
> 10mM
Variations on COSY
DQF-COSY
phase-sensitive experiment
diagonal peaks are narrower
less-sensitive
TOCSY - total correlation spectroscopy
gives correlations for all protons within a spin system
NOESY - Proton-proton through-space interactions via NOE
Gives information about pairs of protons that are close in space (<5 A apart)
Default Parameters
1.5 second relaxation delay
1024 complex points in t2 and 256 increments in t1
2-8 scans per increment
Total time: 1 - 5 hours.
Recommended Concentration
> 10mM
ROESY - rotating frame NOE
Compounds of molecular weight ~1000-2000
Exchange peaks are opposite sign from NOE peaks
HMQC - Heteronuclear Multiple Quantum Correlation experiment
Gives information about strong proton-carbon J-couplings. A strong proton-carbon J-coupling indicates that the proton is directly bonded to the carbon. This experiment gives information that is identical to HETCOR, but because it is proton-detected, it is more sensitive than the standard HETCOR especially on indirect-detection probes.
Default Parameters
1.5 second relaxation delay
512 complex points in t2 and 192 increments in t1
4 scans per increment (no Pulsed Field Gradient) / 1 scan per increment (PFG)
Total time: 20 minutes (no PFG) / 5 minutes (PFG)
Recommended Concentration
> 50mM
HMBC - Heteronuclear Multiple Bond Correlation experiment
Gives information about weak proton-carbon J-couplings. A weak proton-carbon J-coupling indicates that the proton is two, three, or four bonds away from the carbon. This experiment gives information about which protons are near to (but not directly bonded to) different carbons. This experiment (in conjunction with the HMQC) can give an enormous amount of information about molecular structure, since the long range proton-carbon correlations can include quaternary carbons, in addition to protonated carbons.
Default Parameters
1.5 second relaxation delay
512 complex points in t2 and 192 increments in t1
4 scans per increment (no PFG) / 4 scan per increment (PFG)
Total time: 20 minutes (no PFG) / 20 minutes (PFG) (NOTE: PFG HMBC experiments usually have less t1 noise than non-PFG HMBC experiments.)
Recommended Concentration
> 50mM
Carbon-detected experiments
General
Availability
Probes optimized for X-nuclei give better S/N. The direct detection probes are either QNP or Broadband Observe.
Sensitivity is approximately 5700 times less than for proton, so adequate S/N for a reasonable length experiment (10 minutes, 256 scans) requires a concentration of > 100mM.
1D Carbon experiment
Default Parameters
0.8 second acquisition time with proton decoupling
2.0 second relaxation delay (with NOE enhancement)
Spectral width -10ppm to 225ppm
1024-256 scans
1 hour
Recommended Concentration
> 100mM
Special Considerations
Increase d1 for carbons with long T1's (quaternary, carbonyls)
DEPT experiment
Gives information about the number of protons bonded to each carbon.
Default Parameters
0.8 second acquisition time with proton decoupling
2.0 second relaxation delay (with NOE enhancement)
Spectral width -10ppm to 225ppm
96 scans per spectrum (2 spectra total)
Total time: 10 minute
Recommended Concentration
> 100mM
HETCOR - Proton-carbon correlation experiment
Gives information about strong proton-carbon J-couplings. A strong proton-carbon J-coupling indicates that the proton is directly bonded to the carbon. This experiment gives information that is identical to HMQC, but because it is carbon-detected, it is less sensitive than the HMQC experiment.
Default Parameters
2.0 second relaxation delay
512 complex points in t2 and 128 increments in t1
4 scans per increment
Total time: 30 minutes (make a not that time required for HETCOR is approximately 6 times the time required for    HMQC)
Recommended Concentration
> 100mM
Phosphorus-detected experiments
General
Availability
The 500 Bruker Avance, 400 Bruker Avance can run 31P without changing any cable.
Sensitivity is approximately 15 times less than for proton, so adequate S/N for a reasonable length experiment (10 minutes, 256 scans) requires a concentration of > 0.1mM.
1D Phosphorus experiment
Default Parameters
0.8 second acquisition time with proton decoupling
3.0 second relaxation delay (d1; with NOE enhancement)
Spectral width -100ppm to 250ppm
32 scans
Total time: 2 minute
Recommended Concentration
> 0.1mM
Fluorine-detected experiments
General
Availability
The 400 Bruker Avance with QNP probe is best suited for direct observe 19F experiments.
Sensitivity is approximately the same as for proton, so adequate S/N can be obtained with concentration > 0.1mM.
1D Fluorine experiment
Default Parameters
0.8 second acquisition time
3.0 second relaxation delay
Spectral width 150ppm to -200ppm
32 scans
Total time: 2 minute
Recommended Concentration
> 0.1mM
3. 1D Fluorine observe 1H decouple

Setup is exactly like direct observe 19F, only F2 is 1H and uses Waltz 16 decoupling.
为您推荐
您可能想找: 核磁共振(NMR) 询底价
专属顾问快速对接
立即提交
可能感兴趣
icefirel
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
sslin
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 celan 发表:
Fluorine-detected experiments
General Default Parameters
0.8 second acquisition time
3.0 second relaxation delay

How important is the "3.0 second relaxation delay"?
(We used to set up 1 second for D1)
liusky
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
我觉得aq时间是系统自动算出来的。根据你的sw,sequence等等。对于delay时间,可以测一下t1。如果delay时间不够,系统没有充分驰豫,会降低信噪比。