Accurately weigh the 2,4-dinitrophenylhydrazone (2,4 DNP) derivative of the aldehyde to be analyzed into a measured volume of HPLC grade acetonitrile to make a Stock Standard Solution equivalent to 1.0 mg of derivative per ml in acetonitrile. Store the Stock Standard under refrigeration and make fresh monthly. Dilute the Stock Standard Solution in acetonitrile weekly to make 3-5 Working Standards in the range of interest. Store Working Standards in a closed container under refrigeration when not in use. Prior to injection, dilute solution 1:1 with mobile phase, and filter using a 0.45 µm filter.
SAMPLE PREPARATION
Remove each Monitor to be tested from the Return Container. Using forceps or a spatula as a pry, un-snap the Sampling Grid exposing the Sampling Wafer (a single yellow filter disc 3/4" in diameter).
Place the Sampling Wafer into a clean 5 ml glass vial or inert filtration vial (e.g., Whatman Uniprep® Filter Vial, 0.45 µm PVDF, Item No. UN513UAQU). Accurately pipet 1.0 ml of acetonitrile into the vial, close, and seal. Agitate the vial for one minute. Allow the vial to stand 15 minutes, then add 1.0 ml of mobile phase solution, agitate again, and filter the solution through a 0.45 µm filter (or Uniprep). Reserve for HPLC analysis.
BLANK PREPARATION
Remove the Wafer (yellow filter disc) from an unexposed Monitor and process this "BLANK" in exactly the same fashion as the Sample Preparation (above).
HIGH PERFORMANCE LIQUID CHROMATOGRAPHY(HPLC) ANALYSIS
Inject an aliquot of the Sample Preparation from each Monitor to be analyzed into an HPLC System under the following conditions.
HPLC System - Reversed Phase¹ C-18 (15 cm x 3mm id)
Mobile Phase - 50% acetonitrile/50% Acetate Buffer²
(adjust 0.05 M HOAc to pH 5 with KOH in HPLC grade water)
Flow Rate - 1 ml per min (nominal)
Injection Volume - 10-50 microliter (nominal)
Detector Wavelength - 345-365 nm
¹e.g., "Pinnacle TO-11," Restek Corp., Bellefonte, PA ²Conditions given are for formaldehyde. The percentage of acetonitrile may be increased to shorten the elution time for higher aldehydes.
Concomitantly, inject measured aliquots of a BLANK Preparation and three Standard Preparations in the range of interest (i.e. which bracket the concentrations of the Sample Preparations).
CALCULATION
CALCULATION
Determine the Analyte Peak Area as follows:
PEAK AREA (Analyte) =
{ PEAK AREA (Sample Preparation) } - { PEAK AREA(Blank)}
Determine the Analyte Concentration (C) graphically from the PEAK AR versus the Standard Curve prepared by linear regression analysis (least squares).
Calculate Exposure Level from Analyte Concentration as follows.
EXPOSURE LEVEL (ppm) = 1000(C)(V)(F)(R)
(M)(SR)(T)
Where
C = Analyte Concentration Found
(µg/ml Glutaraldehyde di-DNP Derivative)
V = Volume of Solvent (1.0 ml)
F = Fraction of Glutaraldehyde in di-DNP Deriv(0.218)