3.2. Diode laser
AASTunable dye lasers have been proposed as radiation
sources for
AAS already some 20 years ago
w x 38 . However, in spite of their extremely good
spectroscopic properties } they can be set to
virtually any atomic line between 213 nm and 900
nm with a bandwidth corresponding to the natural
width of an atomic line } they have not
found their way into common use because of a
number of practical and economic reasons: these
laser systems are expensive, frequently unreliable,
and difficult to operate. In contrast to dye lasers,
. diode lasers DL appear to be more suitable to
. one day replace hollow cathode lamps HCL and
. electrodeless discharge lamps EDL . Semicon-
ductor laser diodes are nowadays mass produced
for compact disc players, laser printers, optical
data storage systems, and telecommunication
equipment, and hence they are cheap, reliable,
easy to operate and they have long lifetimes. A
number of these diodes have excellent spectroscopic
properties which make them attractive
w x sources for spectrochemical analysis 39 .
Firstly, the power of presently available
commercial DLs is between one and several orders
of magnitude higher than what is provided
by the best HCLs. In addition, DLs show exceptional
stability, both in terms of wavelength and
intensity. These two factors together make it possible
to measure extremely low absorbances if
optimal experimental conditions are realized and
the fundamental shot-noise limit is achieved. But
even in conventional ‘routine’ atomizers, such as
flames and furnaces, detection limits were
achieved that were 1]2 orders of magnitude lower
w x than those obtained with HCLs 39,40 .
Secondly, the typical linewidth of a commercial
DL is approximately two orders of magnitude less
than the width of absorption lines in flames and
furnaces. This makes possible the expansion of
the dynamic range of calibration to high analyte
concentrations by measuring the absorption on
the wings of the absorption line, where optically
w x thin conditions prevail 40 . In addition, a DL,
under normal operating conditions, emits one single
narrow line, which dramatically simplifies the
spectral isolation of the absorption signal. One
does not need the monochromator which is necessary
with HCLs for isolation of the analytical
line from the other spectral lines emitted by the
HCL, and the photomultiplier could be replaced
w x by a simple semiconductor photodiode 39,41 .
Thirdly, the wavelength of DLs can be easily
modulated at frequencies up to GHz by modulation
of the diode current. Wavelength modulation
of the DL with detection of the absorption at the
second harmonic of the modulation frequency, 2f,
. greatly reduces low-frequency flicker noise in
the baseline, providing improved detection limits
w x 39,40 . In addition, wavelength modulation of the
radiation source provides an ideal correction of
non-specific absorption and significantly improves
the selectivity of the
AAS technique.
The major limitation of DL
AAS at this point
in time is that, although a commercial blue laser
diode was introduced earlier this year, the lower
wavelength limit for mass-produced DLs is approximately
630 nm, which means that even with
frequency doubling in non-linear crystals, the important
wavelength range for
AAS of 190]315 nm
cannot be attained yet with this technique. However,
this need not necessarily be a major limitation
for the successful introduction of DL
AASinto routine application, as this technique, in the
opinion of this author, is ideally suited for dedicated
instruments for special purposes and as
low-cost detectors for
gc or HPLC. One example
for such an application is the HPLC]DL
AAS . . system for speciation analysis of Cr III rCr VI
w x proposed by Groll et al. 42 ; another example is
the tungsten coil atomizer DL
AAS system for
the determination of aluminum and chromium
w x described by Krivan et al. 41 , which is shown in
Fig. 5. Another very interesting aspect is that with
DLs as radiation sources, the determination of
non-metals such as halogens, sulfur, or even noble
gases comes within reach of
AAS. All these elements
have long-lived excited states from which
strong absorption transitions can be induced by
w x the red and near-IR radiation of LDs 39 . Zybin
w x et al. 43 , for example reported about a determination
of chlorine by
gc-microwave induced
plasma-DL
AAS with a detection limit some two
orders of magnitude lower than the best values
obtained by optical emission spectrometry.
In the opinion of this author, DL
AAS should
at this point in time not be considered a competitor
of the conventional
AAS with HCLs and
EDLs, but as a very attractive expansion of the
capabilities of
AAS that opens entirely new fields
of application at relatively low cost. The firstgeneration
instruments and modules for DL
AASw x that became commercially available recently 44
offer an excellent opportunity to exploit the potential
of this technique not only in research but
already in routine application.