Determination of emulsion explosives with Span-80 as emulsifier by gas chromatography–mass spectrometry
Fei-Fei Tian, Jing Yu, Jia-Hong Hua, Yong Zhangb, Meng-Xia Xiea
Analytical & Testing Center of Beijing Normal University, Beijing 100875, China
Institute of Beijing Criminal Science and Technology, Beijing 100054, China
a b s t r a c t
A novel approach for identification and determination of emulsion explosives with Span-80 (sorbitol mono-oleate) as the emulsifier and their postblast residues by gas chromatography–mass spectrometry  (gc–MS) has been developed. 24 kinds of emulsion explosives collected have been processed by transesterification  reaction with metholic KOH solution and the emulsifier has turned into methyl esters of fatty acids. From the peak area ratios of their methyl esters, most of these emulsion explosives can be differentiated. In order to detect the postblast residues of emulsion explosives, the sorbitols in the emulsifier Span-80 obtained after transesterification reaction have been further derivatized by silylation reaction with N,O-bis-(trimethylsilyl)trifluoroacetamide (BSTFA) containing 1% trimethylchlorosilane (TMCS) as the derivatizing reagent. The derivatization conditions were optimized and the derivatives were determined by gc–MS. The results showed that the silylation derivatives of sorbitol and it isomers, combined with hydrocarbon compounds and methyl esters of fatty acids, were the characteristic components for identification of the emulsion explosives. The established approach was applied to analyze the postblast residues of emulsion explosives. It has been found that the method was sensitive and specific, especially when detecting the derivatives of sorbitol and its isomers by gc–MS in selecting ion mode. The information of the characteristic components can help probe the origin of the emulsion explosives and providing scientific evidences and clues for solving the crimes of the emulsion explosive explosion.
Keywords: Emulsion explosives  Span-80    Postblast residues  Derivatization  gc–MS
概述:
建立一种采用gc-MS法对乳化炸药(乳化剂为Span-80)及其爆炸残留物鉴定和识别的分析方法。
本文收集到的24类乳化炸药,在氢氧化钾的甲醇溶液中进行酯交换反应生成脂肪酸甲酯,根据得到的脂肪酸甲酯峰面积相对比值的差异,实现24类乳化炸药种类识别。为了检测乳化炸药的爆炸残留物,对乳化炸药的另一酯交换产物山梨醇进行硅烷化衍生处理。实验对衍生化条件进行优化并且利用gc-MS法对衍生产物进行分析,结果表明:山梨醇及其同分异构体的硅烷化衍生产物,脂肪酸甲酯和碳氢组分是鉴定乳化炸药及其爆炸残留物的特征成分;利用gc-MS-SIM法来测定乳化炸药爆炸残留物中的山梨醇及同分异构体具有更好的专属性和更高的灵敏度。乳化炸药及其爆炸残留物的特征成分可以为探索乳化炸药的来源和抓捕乳化炸药爆炸犯罪嫌疑人提供科学的线索和依据。
1Introduction
Identification and determination of explosives, especially their post blast residues, was an important topic in forensic science[1,2]. Determining the chemical compositions of explosives and their post blast residues can provide scientific information for differentiation of these explosives and probing their origins, and offer important clues and evidences for solving the crimes of the explosion.
The analysis of explosives and explosive residues has attracted the attention of the scientific community in recent decade, and various methods have been developed. Ion chromatography[2-4] with conductivity detector by connecting anion-exchange and cation-exchange columns has been successfully used to detect the inorganic ions (anions and cations) in inorganic explosive residues. Capillary electrophoresis (CE) approaches has been applied to detect both the inorganic ion[5-9] and the organic components, such as nitroaromatic[10-12] and benzoate ions[13] in the explosive samples.
The postblast explosive residues can be extracted and enriched with solid-phase microextraction procedures and then determined by gas chromatography (gc)[14] with nitrogen phosphorus[15] or electron capture detectors[16-19]. gc coupled to mass spectrometry (MS) was also a powerful method for characterization and identification of explosives and explosive residues[20-25], especially combined with other approaches.
High-performance liquid chromatographic methods[26,27] and various technologies of mass spectrometry have been attempted to determine the explosive residues recent years. Electrospray ionization (ESI) and atmospheric pressure chemical ionization(APCI) mass spectrometry[28,29] can identify explosive oxidizers[30], stabilizers[31] and explosives residues. The composition of explosives was detected by electrospray ionization Fourier transform ion cyclotron resonance mass spectrometry[32]. Isotope ratio mass spectrometry (IRMS)[33], ion mobility spectrometry (IMS)[34-36] and laser electrospray mass spectrometry[37,38] can also examine the organic explosives and their residues. Some new mass spectrometric devices have been designed to detect trace explosives on various surfaces[39-41].
Fourier transform infrared spectroscopy (FT-IR)[42,43] , laser induced breakdown spectroscopy[44-47], chip-based isotachophoresis[48] and fluorescent-labeled imprinted polymer[49] have also been developed for determination of explosive residues.
Recent years, emulsion explosives are most widely used and their annual output was about 40 to 50 % for all the explosives. Emulsion explosives comprise two immiscible phases, an aqueous solution of inorganic oxidizing salt as a discontinuous phase dispersed throughout a continuous organic fuel phase, and emulsifiers are used to improve the stability of emulsion explosive composition[50]. The commonly used emulsifier is sorbitan mono-oleate (Span-80) and polyisobutylene succinimide (PIBSA) and its amount is generally within a range from 0.1 to 5.0%.
There are few publications concerning the identification of emulsion explosives[51,52]. Korosec RC determined the content of ammonium nitrate and sodium nitrate in explosive using thermogravimetry (TG) and differential scanning calorimetry (DSC) methods. Tata D. and coauthors reported a liquid chromatography-mass spectrometry method for characterization of the emulsifiers, both Span-80 and PIBSA, while only Span-80 was detected in postblast residues of emulsion explosives.
In this report, the identification and determination approaches for emulsion explosives with span-80 as emulsifier and their postblast residues have been systematically investigated based on gc/MS in full scan and selected ion monitoring modes. The samples were processed with transesterification and silylation reactions, and the derivatizing conditions were optimized. The established method was sensitive, repeatable and reliable for detecting the emulsifier in the postblast residues of emulsion explosives, and the chromatographic profiles of the derivatives for sorbitol and its isomers can give rich information to probe the origin of the emulsion explosives.
2. Experimental and methods
2.1. Reagents and preparation of solutions
  N,O-bis(trimethylsilyl) trifluoroacetamide (BSTFA) + trimethylchlorosilane (TMCS) (BSTFA:TMCS=99:1) were obtained from sigma-Aldrich company (MO. USA). Methanol and hexane were HPLC grade and purchased from Dikma technology (CA, USA). The emulsifier of industrial pure Span-80 (95%) was purchased from Beijing shiji Chemical (Beijing, China). Other reagents were analytical grade and from Beijing chemical reagent factory (Beijing, China). All the reagents were used as supplied without further purification.
Emulsifier samples (span-80) were accurately weighted (100 mg) and dissolved in a 100 ml volumetric flask with hexane. The working solutions were prepared by diluting the stock solution in hexane as required.
Methanolic KOH solution was prepared by dissolving 2.8 g KOH with 50 ml anhydrous methanol. The required concentrations can be obtained by diluting the solution with methanol.
2.2. Gas chromatography/mass spectrometry
The gc/MS analyses were performed using a Trace gc-MS spectrometer (Thermofisher, CA, USA) equipped with a DB-5MS fused-silica capillary column (30 m × 0.25 mm I.D., 0.25μm film thickness) which was from J & W Scientific (Folsom, CA, USA). Data was collected with an Xcalib software data process system. Complete characterization of the components was carried out in full scan mode and in selected ion monitoring mode.
The gc/MS analyses were carried out in splitless mode using high purity helium as carrier gas at a flow rate of 1 ml/min. The ion source temperature was 200 ºC and the gc-MS interface was 250 ºC. The injection port temperature was 250ºC; the oven was maintained at an initial temperature of 60 ºC for 2 min, and then programmed at 10 ºC /min to a final temperature of 290 ºC where it was maintained for 10 min.
2.3 Derivatization reactions
2.3.1 Transesterification reaction for the emulsifier Span-80 and emulsion explosives
A procedure: the 10 mg emulsion explosive was extracted by hexane (10 mL) with two times, combined the extraction solutions and filtered using a 0.45-μm nylon filter for removal of insoluble particles. The solution was then concentrated to 1 mL with a gentle stream of nitrogen. Add 0.1 mL 0.1 mol/mL methanolic potassium hydroxide solution to the concentrated extraction solution, and then transesterification reaction was carried out in ultrasonic bath for 15 min at room temperature.
B procedure: the 1.0 mL SP-80 hexane solution (100 μg/mL) was added with 0.1 mL 0.1 mol/mL methanolic potassium hydroxide solution, and then reacted in ultrasonic bath for 15 min at room temperature.
C procedure: the above reaction mixtures were transferred to a separatory funnel, respectively. After addition 5 mL hexane and 5 mL sodium sulfate saturated aqueous solution, the two phases were separated. The hexane layer containing the methyl esters was washed another 5 mL sodium sulfate saturated water solution. The hexane solution was dried with anhydrous sodium sulfate and vaporized to near dryness with a gentle stream of nitrogen. The residues were dissolved by 1.0 mL hexane and then analyzed by gc/MS instrument.
2.3.2. Acetylation and silylation reactions for the emulsifier SP-80 in explosive
Acetylation reaction:  the reaction mixture obtained in B procedure was dried with nitrogen gas. Add 0.5 mL pyridine and 0.5 mL acetic anhydride to the test tube of residues, vortexed to obtain a uniform solution. Seal the tube with its plug and put the tube to a water bath with 90 ˚С for 40 min. The reaction mixture was dried with nitrogen gas and the residue was extracted with 2.0 mL dichloromethane, and the extraction solution was washed with saturated NaCl water solution. The dichloromethane layer was separated and dried with anhydrous sodium sulfate and then was analyzed by gc/MS.
Silylation reaction: the reaction mixtures obtained in A and B procedures were vaporized with nitrogen, individually. Add 0.5 mL pyridine and 0.2 mL BSTFA and TMCS (v/v 99:1), vortexed and sealed the tube, and then put the tube to a water bath with 60 ˚С for 40 min. The reaction mixture was dried with nitrogen gas and the residue was dissolved with 1.0 mL hexane before gc/MS analysis.
2.4. Collection of samples and Preparation of Post blast Residues
Tweety four emulsion explosives were collected from different manufacturers in various restricts of China.
Steel plates (200 cm × 100 cm, thickness 100 μm) were used to collect explosive residues from field test and were placed on the ground. The explosive device was positioned in the center of the plate. Four further plates were arranged laterally around the explosives at a distance of 1-2 m. The emulsion explosive was detonated and the residues on the steel plates were collected in polyethylene bags, which were subsequently heat-sealed to avoid contamination. Soil samples were also taken directly near the base plate both before and after the detonation [7].
The residues on each steel plate were washed with 30 mL hexane. The hexane solutions were combined and filtered using a 0.45-μm nylon filter. The solution was concentrated by rotary evaporator to about 5.0 mL and then transferred to a test tube. The solution in the tube was further concentrated to about 1.0 mL with nitrogen gas.
The soil samples (about 20 g) were extracted with 30 mL hexane twice, and the combined solution was concentrated following above procedures.
The solutions of residues were derivatized according to above procedures of B and Silylation reaction before gc/MS analysis.
3.  Results and discussion
3.1 Differentiation of emulsion explosives with Span-80 as Emulsifier
3.1.1 Separation of emulsion explosives by gc-MS
The compositions of emulsion explosives mainly include ammonium nitrate, fuels, emulsifiers, and other organic or inorganic additives[50], and hydrocarbons are the commonly used fuel sources.
Twenty four samples of emulsion explosives with Span-80 as emulsifier were extracted with hexane individually and the extracts were analyzed by gc-MS based on the chromatographic conditions in experimental section. The results showed that the volatile ingredients of the emulsion explosives were n-alkane with carbon number distribution ranged from C21 to C33 (Figure not showed). The reproducibility of the separation was satisfactory and the n-alkane carbon number distribution remained constant with various samples. There is no other information about the compositions of the emulsion explosives in the chromatograms, and it is difficult to achieve individual differentiation of these explosives due to similar chromatographic features of their hydrocarbon components.
3.1.2 Differentiation of emulsion explosives
Emulsifiers are the critical components of emulsion explosives, and the industry Span-80 is a complex mixture. It contains not only the product of a single ester and its isomer, but also other pairs of esters, multi-esters and their corresponding isomers[53]. The compositions of Span-80 from various origins would be different, and this provides a clue to differentiation of the emulsion explosives.
Span-80 is a nonionic surfactant and can not be measured directly by gc-MS due to its non-volatile. According to the structural characteristics of Span-80, transesterification reaction was performed[54,55] in metholic KOH solution, and the fatty acids in Sp-80 molecules were dissociated and transformed into methyl esters. Fig.1 showed the gc/MS total ion chromatogram of Span-80 after the transesterification reaction. The three main peaks in the chromatogram were determined by their MS spectra, which were methyl oleate, methyl linoleate and methyl palmitate respectively. The boiling point of oleic acid was nearly same as that of linoleic acid, and they were difficult to be separated by rectification method, and so their mixtures were usually used to industrial synthesis of Span-80[56,57], while palmitinic acid was an impurity component of the oleic acid. The ratio of each component for these fatty acids would be different from product to product, and these differences can be utilized to differentiate the Span-80, and further to distinguishes the emulsion explosives.
Fig. 1 The gc/MS total ion chromatogram of Span-80 after transesterification reaction,
1: methyl palmitate; 2: methyl linoleate; 3-4: methyl oleate; 5: methyl stearate.
Twenty-four kinds of the industrial emulsion explosives were extracted with hexane and derivatized by the transesterification reaction as described in experimental section, and the derivatives were analyzed by gc-MS. The results showed that the relative peak area for the methyl ester of oleic acid varied from 23.7% to 97.2%, while those of linoleic acids from 73.6% to 1.2%, and the percentages of impurity, palmitinic acid, were below 10% for most of the emulsion explosives. The relative standard deviations (RSD) of 5 repeats for their relative peak areas were below 2%.
Figure 2 is the bar chart presenting the relative peak area of the methyl esters of oleic acid and linoleic acid for the 24 explosives. From the Figure, it can be seen that the most of the 24 explosive samples can be differentiated according to the relative intensities of the fatty acids in the derivatives of their emulsifier. It should be noted that the quality of span-80 was directly proportional to the percentage of oleic acid and lower content of oleic acid led to poor quality of the emulsifier. The results in Fig.2 showed that the percentage of oleic acid in the emulsifier of some explosives was lower than 50%, and these explosives may be suitable to industrial pueposes.
Fig.2 The relative peak area of the methyl esters of oleic acid versus that of linoleic acid for the 24 industrial emulsion explosives.
3.2 Derivatization of the emulsifier
The postblast residues of emulsion explosives were a complex matrix, and there have been many interferential components on the site of explosion[43]. As described above, the industrial emulsion explosives can be determined and differentiated by the presence and the ratios of the fatty acids in their emulsifier span-80. While in the postblast residues, the fatty acids may sometimes exist in the environment, such as soil, the cotton for collecting samples, and so the fatty acids were not the characteristic components to signify the presence of emulsifier span-80.
From the compositions of span-80, it can be seen that sorbitols are characteristic components, and they seldom exist in environment. If sorbitols have been detected from the postblast residues of explosives, it can be concluded that the explosives were emulsion explosives with Span-80 as emulsifier.
The sorbitol esters in Span-80 are non-volatile and can not be analyzed directly by gc/MS and so derivatization reaction is needed. Direct silylation of Span-80 could improve their volatilities, and their derivatives can be separated by gc method[58]. This approach can be utilized to monitor the industrial products of Span-80, while the limits of detection were not enough to analyze the emulsifier in postblast residues of emulsion explosives.
3.2.1 Selection of derivatization reagent for sorbitols
In Span-80, sorbitols exist in various forms, 1,4:3,6-isosorbide, 1,4-sorbitan, 1,5-sorbitan and other isomers. As shown in Fig.3, after the transesterification reaction, the Span-80 turned into the methyl esters of fatty acids and the isomers of sorbitols which should be further derivatized before gc/MS analysis. Acetylation and silylation are usually utilized to convert alcohols into volatilizable derivatives[59-61]. In current work, the common derivatising agents like acetic anhydride–pyridine for acetylation and N,O-bis-(trimethylsilyl)trifluoroacetamide (BSTFA) containing1% trimethyl- chlorosilane (TMCS) for silylation were investigated.
Fig.3 The derivatization reactions of Span-80
The Span-80 (100 μg/ml) was derivatized with acetic anhydride and BSTFA individually after transesterification reaction as described above. The derivatives were separated by gc/MS, and their components were qualitatively determined by search from the NIST08 data base which have about the mass spectra of 180,000 compounds and confirmed by the ion fragments of their mass spectra.
Fig.4 showed the total ion chromatograms of the derivatives. From Fig. 4a, it can be seen that only the acetylation derivative of 1,4-sorbitan can be detected apart from the methyl esters of fatty acids, while the signal of the derivative was very weak. From the mass spectra of the acetylation derivatives (Figure was not shown), the main ion fragments of the derivatives were below m/z 100, and it indicated that acetylation reaction for the sorbitols was not suitable to monitor the postblast vestiges of the emulsion explosives. Fig 4-b showed the total ion chromatograms for the silylation derivatives of Span-80 after transesterification reaction. From the search results and their mass spectra, the derivatives of these components were determined.
Fig.4a The gc–MS total ion chromatogram of Span-80 (100 μg/ml) after transesterification and acetylation reactions. Peak:1acetylated 1,4-sorbitan; (2) methyl palmitate; (3) acetylated sorbitol; (4) methyl linoleate; (5-6) methyl oleate; (7) methyl stearate.
Fig.4b The gc–MS total ion chromatogram Span-80 (100 μg/ml) after transesterification and silylation reactions. Peak: (1) 1,4:3,6-isosorbide-TMS; (2) 1,4-sorbitan-TMS; (3-7) unspecified isomers of cyclized sorbitol-TMS; (8-10) uncyclized sorbitol-TMS; (11) Palmatic acid –TMS; (12) methyl linoleate; (13-14) methyl oleate; (15) methyl stearate; (16) Linoleic acid-TMS; (17) Oleic acid-TMS; (18) Stearic acid-TMS.
Table 1 listed the search results of the derivatives and the main ion fragments in their mass spectra apart from the fragments at m/z 73, 117, 147 of the derivative reagent and those fragments (m/z below 100). From Table 1, it can be seen that peak 1 to peak 10 were the trimethylsilyl (TMS) of sorbitols and its various isomers, and other components were methyl ester or silylation derivatives of fatty acids. The intensities of these components were significant and the results illustrated that silylation of the emulsifier after transesterification reaction can give rich information of its characteristic constituents. Otherwise, the main ion fragments of the derivatives were in high mass range, which would be favor to detect the components in selected ion monitoring mode (SIM) by gc/MS.
Table 1 Main ion fragments and search results for the silylation derivatives of Span-80
Peak number
Molecular
weight
Main ion fragments
Name of components
1
290
101(70); 143(30); 157(30); 204(25); 275(20); 290(M+,5)
1,4:3,6-isosorbide-(TMS)2
2
452
103(80); 169(30); 217(100); 259(55); 272(80); 362(10)
1,4-sorbitan-(TMS)4
3
452
103(40); 192(30); 205(40); 217(35); 259(40); 362(25)
cyclized sorbitol-(TMS)4
4
452
103(40);157(20);192(30);205(40);217(35);259(35);362(30)
5
452
103(30); 192(15); 205(20); 217(55); 259(20); 362(10)
6
452
103(70);192(80);205(55);217(80);259(60); 362(20)
7
452
103(55);157(40); ;205(50); 217(50); 259(45); 362(30).
8
614
103(30); 205(40); 217(35); 307(30); 319(55)
sorbitol-(TMS)6
9
614
103(60); 205(75); 217(70); 307(40); 319(65)
sorbitol-(TMS)6
10
614
103(70); 205(85); 217(60); 307(35); 319(60)
sorbitol-(TMS)6
11
270
129(20); 143(50); 227(25); 270(M+,30)
methyl palmitate
12
294
109(35); 164(30); 262(25); 294(M+,15)
methyl linoleate
13
296
109(45);130(40);180(30); 264(35);296(M+,10)
methyl oleate
14
296
109(30);130(20);180(20); 264(30);296(M+,10)
isomer of methyl oleate
15
298
74(100);101(95);143(30);255(15);298(M+,20)
methyl stearate
16
352
110(20);129(35);150(20);337(30);352(M+,5)
linoleic acid-TMS
17
354
129(55);145(30);339(55);354(M+,10)
oleic acid-TMS
18
356
129(45);132(75);145(40);341(70);356(M+,15)
stearic acid-TMS
TMS: trimethylsilyl
3.2.2 Optimization of the derivatization conditions
To guarantee the repeatability and reliability of the results, the conditions of the derivatization reactions have been systematically investigated and several related factors, such as the concentration of the methanolic KOH, the amount of derivatization reagent, the reaction temperature and time duration of the reaction have been optimized.
The transesterification reaction is usually carried out in alkali solution at ambient temperature, such as methanolic KOH solution. The concentrations of the methanolic KOH from 0.1mol/L to 0.8 mol/L were tested to monitor their influence on the following silylation reaction. The time duration of the silylation derivatization reaction was 60 min and a large excess of the derivatization reagent (500μL) was utilized to assure the completeness of the derivatization reaction. The peak areas of sorbitols-TMS derivatives versus the concentration of methanolic KOH were shown in Fig. 5.
Fig.5 Relationship between the peak area of sorbitols derivatives and the concentration of methanolic KOH, the concentration of Span-80 was 100 μg/ml. line A: the total peak area of sorbitols-TMS derivatives; line B: the peak area of peak 2 to peak 7; line C: the peak area of 1,4:3,6-isosorbide-TMS.
From Fig.5, it can be seen that the peak area of the sorbitol trimethylsilyl derivatives remained constant when the concentration of methanolic KOH was below 0.2 mol/L, while the peak area gradually decreased with increasing the alkali concentration. In higher alkali concentration, Span-80 would be saponified in some extents, which may hinder the silylation reaction of the sorbitols. Therefore, the concentration of the methanolic KOH was fixed at 0.1 mol/L.
The reaction was conducted under 25, 40, 60, 90 ˚С, individually, to investigate the influence of temperature on the derivatization reaction and the results were shown in Fig.6. The results indicated that the peak area of the sorbitol derivatives reached constant when the temperature was above 40 ˚С. In order to increase the reaction rate, the derivatizing reaction was performed at 60 ˚С.
Fig.6 Relationship between the peak area of sorbitols derivatives and the reaction temperature, other conditions were same as those in Fig.5.
The amount of BSTFA and the duration of reaction time on the peak area of sorbitols-TMS derivatives were also optimized. 100 μL, 200 μL, 300 μL, 400 μL, 500 μL derivative reagent were tested, respectively. The results showed that the derivatization reaction can be conducted completely when the amount of derivative reagent was above 200 μL. The duration of reaction ranged from 20 min to 80 min was tried and the results showed that the peak area of sorbitol derivatives can be reached constant with above optimized conditions when the reaction time was 40 min.
In conclusion, the optimum conditions for the derivatization reaction were as follow: the concentration of the methanolic KOH for transesterification reaction of Span-80 was 0.1 mol/L; the silylation reaction was conducted under 60 ˚С for 40 min with 200 μL derivatization reagent of BSTFA. The sorbitol derivatives were stable, and the relative standard deviations (RSD) of their peak area within 10 h were below 3%.
Based on the derivatizing procedures established above, five different concentrations of Span-80 (10, 25, 50, 100, 200 μg/ml) were derivatized and analyzed by gc/MS in full scan mode. The linearity between the peak areas of the sorbitol derivatives and the corresponding concentrations of Span-80 was satisfactory, and the line regression coefficient was 0.995 in the concentration range from 10 to 200 μg/ml. The limits of detection for Span-80 were 1.0 μg/ml according to the signal-to-noise ratio of 3. The results indicated that the established approach has high sensitivity and with satisfactory repeatability, and can be utilized to detect the postblast residues of the emulsion explosives.
3.3. Analysis of emulsion explosives
The analytical procedure established above was used to analyze the characteristic components of emulsion explosives, and Fig.7 showed the full scan gc/MS chromatogram for emulsion explosives after transesterification and silylation reactions. In Fig.7, the components A, B and C were the silyl derivatives of sorbitol and its isomers; components D were the derivatives of fatty acids, and components E were hydrocarbon compounds. These characteristic components can determine the emulsion explosives.
Fig.7 The gc-MS total ion chromatogram of the emulsion explosive after derivatization reactions, the explosives was 10 mg, component A: 1,4:3,6-isosorbide-TMS; B: isomers of cyclized sorbitol-TMS; C: uncyclized sorbitol-TMS; D: methyl esters of fatty acids; E: n-alkane with carbon number distribution ranged from C21 to C33.
In order to improve the sensitivity for detecting sorbitols and its isomers, selected ion monitoring mode (SIM) was employed in gc/MS analysis. The ions selected for trimethylsilyl (TMS) derivative of 1,4:3,6-isosorbide were 290(M+), 275(M-15+, loss of a -CH3) and 157 (M-133+, loss of –O-Si(CH3)3 and fragment –CH2-CH2-O- ). The mass spectra of components B were similar, and so the ions selected in this time window were same. 362(M-90+, loss of (CH3)3SiO- and H), 259 (M-193+, loss of (CH3)3SiO-, (CH3)3SiO-CH2- and H) ion and 217 (M-235+, loss of (CH3)3SiO-, (CH3)3SiO-CH2- and fragment –CH-CH2-O- ); For components C, the selected ions were 319((M-295+, loss of 2(CH3)3SiOCH2- and (CH3)3SiO-), 217(M-397+, loss of 3(CH3)3SiO- and (CH3)3SiOCH2- and fragment –CH-CH2-), 205( ion fragment (CH3)3SiO-CH+-CH2-SiO(CH3)3 ).
Fig.8 showed the gc/MS chromatogram of sorbitol derivatives in SIM mode. It can be seen from Fig.8 that the signals of the derivatives were significant and the method established was sensitive and can be utilized to monitor sorbitols in the postblast residues of emulsion explosives.
Fig. 8 The gc-MS selected ion chromatogram of the emulsion explosive after derivatization reaction, the explosive was 10 mg, peak:(1) 1,4:3,6-isosorbide-TMS; (2-7) 1,4-sorbitan-TMS and its isomers; (8) sorbitol-TMS
The hydrocarbon compounds in Fig.7 were normal alkyl paraffins and their chromatogram can be further processed by extracted ion chromatogram (m/z 85) [52] (see Fig.9). The extracted ion chromatogram of the hydrocarbon compounds can clearly display characteristic profiles of these components.
Fig.9 Extracted ion chromatogram (m/z 85) of the hydrocarbon compounds in the emulsion explosive
3.4  Postblast Residue Analysis of emulsion explosives
The described method was applied to identification for the post-blast residues of emulsion explosives. The residues were collected and pretreated with transesterification and silylation reactions, and then analyzed by gc/MS in full scan mode. Fig.10 showed the total ion chromatogram for the postblast residues of the emulsion explosive. It can be seen from the profile of Fig.12 that the characteristic components of the emulsion explosive (components A to E) remained in the residues.
Fig.10 The total ion gc/MS chromatogram in full scan mode for the postblast emulsion explosive residues after derivatization reactions. components A: 1,4:3,6-isosorbide-TMS; B: isomers of cyclized sorbitol-TMS; C: uncyclized sorbitol-TMS; D: fatty acid methyl ester; E: n-alkane with carbon number distribution ranged from C16 to C31. The emulsion explosive for explosion was 300 g.
The chromatogram profiles of sorbitol components would become clearer when the postblast residues of the emulsion explosives were monitored by gc/MS in SIM mode (see Fig. 11). As shown in Fig.11, the sorbital components for postblast of emulsion explosive remained comparing with that for the original emulsion explosive, while profiles of the components have been changed. Blank measurements (the soil in the site of explosion) were performed in the same method and the results indicated that there was no interference for identification of the sorbitol and its isomers. The information obtained could be used for determining the origin and kinds of the emulsion explosives.
Fig. 11 The selected ion gc/MS chromatogram of the postblast explosive residues.
Peak (1):1,4:3,6-isosorbide-TMS; (2-7):1,4-sorbitan-TMS and its isomers; (8): sorbitol-TMS
Fig.12 Extracted ion chromatogram (m/z 85) of the hydrocarbon compounds for the postblast emulsion explosive residues
Fig.12 showed the extracted ion chromatogram (m/z 85) of the hydrocarbon components in the postblast residues of the emulsion explosive. It can be seen that the carbon number distribution of hydrocarbon components was between C15 and C31, which have some differences with that of preblast emulsion explosive (see Fig.9, which was between C21 and C33 ). It showed that the hydrocarbon components underwent degradation during the process of explosion. The chromatogram profiles of the hydrocarbon components appeared as Gauss distribution. The experimental results also indicated that the carbon number distribution of hydrocarbon components changed into C14 and C25 when the emulsion explosive was increased to 600 g. The phenomena were similar with those[52] previously reported.
4 Summary and Conclusions
Twenty four kinds of emulsion explosives with Span-80 as emulsifier were collected and analyzed by gc/MS in full scan and SIM modes. The emulsifier Span-80 was not volatile, and the gc/MS chromatograms for emulsion explosives can only give the information of hydrocarbon components. After transesterification reaction in the methanolic KOH solution, Span-80 can be turned into sorbitols and methy esters of fatty acids, which can be analyzed by gc/MS. The presence of emulsifier derivatives in emulsion explosives serves as a useful marker to distinguish from other inorganic or organic explosives. From the ratios of different fatty acid methyl esters, most of the emulsion explosives collected can be individually identified.
In order to establish an approach for determining the post-blast residues of emulsion explosives, the emulsifier Span-80 was derivatized with BSTFA after transesterification reaction, and the derivatizing conditions were optimized. The derivatives were separated by gc/MS method, and the components of sorbitols were measured from their mass spectra. The established method was applied for analysis the post-blast residues of emulsion explosives, and several classes of characteristic components, including sorbitol and its isomers, fatty acid methyl esters, and hydrocarbon compounds, have been detected. From the chromatographic profiles of these components, the kinds of emulsion explosives could be inferred.
The results showed that the established approach has high sensitivity and repeatability for detecting the post blast residues of emulsion explosives and it can also avoid the interferences of environmental substances.
5 References
[1]          D.S. Moore, Rev. Sci. Instrum. 75 (2004) 2499.
[2]          C. Johns, R.A. Shellie, O.G. Potter, J.W. O'Reilly, J.P. Hutchinson, R.M. Guijt, M.C. Breadmore, E.F. Hilder, G.W. Dicinoski, P.R. Haddad, J. Chromatogr. A 1182 (2008) 205.
[3]          G.W. Dicinoski, R.A. Shellie, P.R. Haddad, Anal. Lett. 39 (2006) 639.
[4]          H.B. Meng, T.R. Wang, B.Y. Guo, Y. Hashi, C.X. Guo, J.M. Lin, Talanta 76 (2008) 241.
[5]          C. Johns, M.C. Breadmore, M. Macka, M. Ryvolova, P.R. Haddad, Electrophoresis 30 (2009) S53.
[6]          M. Pumera, Electrophoresis 27 (2006) 244.
[7]          J.P. Hutchinson, C.J. Evenhuis, C. Johns, A.A. Kazarian, M.C. Breadmore, M. Macka, E.F. Hilder, R.M. Guijt, G.W. Dicinoski, P.R. Haddad, Anal. Chem. 79 (2007) 7005.
[8]          J.P. Hutchinson, C. Johns, M.C. Breadmore, E.F. Hilder, R.M. Guijt, C. Lennard, G. Dicinoski, P.R. Haddad, Electrophoresis 29 (2008) 4593.
[9]          T. Kishi, J. Nakamura, H. Arai, Electrophoresis 19 (1998) 3.
[10]        B.C. Giordano, D.S. Burgi, G.E. Collins, J. Chromatogr. A 1217 (2010) 4487.
[11]        S.A. Oehrle, J. Chromatogr. A 745 (1996) 233.
[12]        Y.Y. Yang, J.T. Liu, C.H. Lin, Electrophoresis 30 (2009) 1084.
[13]        D.J. Klapec, D. Ng, J. Forensic Sci. 46 (2001) 1168.
[14]        H.T. Mayfield, E. Burr, M. Cantrell, Anal. Lett. 39 (2006) 1463.
[15]        G.L. Burleson, B. Gonzalez, K. Simons, J.C.C. Yu, J. Chromatogr. A 1216 (2009) 4679.
[16]        R. Waddell, D.E. Dale, M. Monagle, S.A. Smith, J. Chromatogr. A 1062 (2005) 125.
[17]        O.L. Collin, C. Niegel, K.E. DeRhodes, B.R. McCord, G.P. Jackson, J. Forensic Sci. 51 (2006) 815.
[18]        Y.P. Huang, Y.C. Yang, Y.Y. Shu, J. Chromatogr. A 1140 (2007) 35.
[19]        S. Calderara, D. Gardebas, F. Martinez, Forensic Sci. Int. 137 (2003) 6.
[20]        M. Berg, J. Bolotin, T.B. Hofstetter, Anal. Chem. 79 (2007) 2386.
[21]        S. Jonsson, L. Gustavsson, B. van Bavel, J. Chromatogr. A 1164 (2007) 65.
[22]        K.S. Bradley, J. Forensic Sci. 50 (2005) 96.
[23]        S.P.SharmaS.C.Lahiri, J. Energetic Mat. 23 (2005) 239.
[24]        D.K.Kuila, S.C.Lahiri, J. Indian Chem. Soc. 84 (2007) 74.
[25]        J.M. Perr, K.G. Furton, J.R. Almirall, Talanta 67 (2005) 430.
[26]        S. Babaee, A. Beiraghi, Anal. Chim. Acta 662 (2010) 9.
[27]        D. Gaurav, A.K. Malik, P.K. Rai, Crit. Rev. Anal. Chem. 37 (2007) 227.
[28]        I. Cotte-Rodriguez, H. Hernandez-Soto, H. Chen, R.G. Cooks, Anal. Chem. 80 (2008) 1512.
[29]        E. Holmgren, H. Carlsson, P. Goede, C. Crescenzi, J. Chromatogr. A 1099 (2005) 127.
[30]        X.M. Zhao, J. Yinon, Rapid Commun. Mass Spectrom. 16 (2002) 1137.
[31]        D. Perret, S. Marchese, A. Gentili, R. Curini, A. Terracciano, E. Bafile, F. Romolo, Chromatographia 68 (2008) 517.
[32]        Z.G. Wu, C.L. Hendrickson, R.P. Rodgers, A.G. Marshall, Anal. Chem. 74 (2002) 1879.
[33]        S.J. Benson, C.J. Lennard, P. Maynard, D.M. Hill, A.S. Andrew, C. Roux, Sci. Justice 49 (2009) 73.
[34]        M. Tabrizchi, V. ILbeigi, J. Hazard. Mater. 176 (2010) 692.
[35]        H. Koyuncu, E. Seven, A. Calimli, TUrk. J. Chem. 29 (2005) 255.
[36]        T.L. Buxton, P.D. Harrington, Appl. Spectrosc. 57 (2003) 223.
[37]        J.J. Brady, E.J. Judge, R.J. Levis, Rapid Commun. Mass Spectrom. 24 (2010) 1659.
[38]        Y.Y. Mou, J.W. Rabalais, J. Forensic Sci. 54 (2009) 846.
[39]        N.L. Sanders, S. Kothari, G.M. Huang, G. Salazar, R.G. Cooks, Anal. Chem. 82 (2010) 5313.
[40]        H.W. Gu, S.P. Yang, J.Q. Li, B. Hu, H.W. Chen, L.L. Zhang, Q. Fei, Analyst 135 (2010) 779.
[41]        Y. Zhang, X.X. Ma, S.C. Zhang, C.D. Yang, Z. Ouyang, X.R. Zhang, Analyst 134 (2009) 176.
[42]        K. Banas, A. Banas, H.O. Moser, M. Bahou, W. Li, P. Yang, M. Cholewa, S.K. Lim, Anal. Chem. 82 (2010) 3038.
[43]        A. Banas, K. Banas, M. Bahou, H.O. Moser, L. Wen, P. Yang, Z.J. Li, M. Cholewa, S.K. Lim, C.H. Lim, Vib. Spectrosc 51 (2009) 168.
[44]        V. Lazic, A. Palucci, S. Jovicevic, C. Poggi, E. Buono, Spectroc. Acta Pt. B-Atom. Spectr. y 64 (2009) 1028.
[45]        J.L. Gottfried, F.C. De Lucia, C.A. Munson, A.W. Miziolek, Anal. Bioanal.Chem. 395 (2009) 283.
[46]        J.L. Gottfried, F.C. De Lucia, A.W. Miziolek, J. Anal. At. Spectrom. 24 (2009) 288.
[47]        F.C. De Lucia, J.L. Gottfried, A.W. Miziolek, Opt. Express 17 (2009) 419.
[48]        J.E. Prest, M.S. Beardah, S.J. Baldock, S.P. Doyle, P.R. Fielden, N.J. Goddard, B.J.T. Brown, J. Chromatogr. A 1195 (2008) 157.
[49]        R.C. Stringer, S. Gangopadhyay, S.A. Grant, Anal. Chem. 82 (2010) 4015.
[50]        H.A.Bampfied, J.Cooper, Encyclopedia of Emulsion Technology, vol.3, Ed. P.Becher, New York1998,282-292.
[51]        R.C. Korosec, P. Kajic, P. Bukovec, J. Therm. Anal. Calorim. 89 (2007) 619.
[52]        D. De Tata, P. Collins, N. Campbell, J. Forensic Sci. 51 (2006) 303.
[53]        Y.J. Wang, X.G. Wang, S.L. Yan, J. Univ. Sci. Technol. Beijing 13 (2006) 102.
[54]        U. Rashid, F. Anwar, B.R. Moser, S. Ashraf, Biomass Bioenergy 32 (2008) 1202.
[55]        M.R. Sahasrabudhe, R.K. Chadha, J. Am. Oil Chem. Soc. 46 (1969) 8.
[56]        C.F. Torres, L.P. Lessard, C.G. Hill, Biotechnol. Progr. 19 (2003) 1255.
[57]        R. Gulati, P. Arya, B. Malhotra, A.K. Prasad, R.K. Saxena, J. Kumar, A.C. Watterson, V.S. Parmar, Arkivoc (2003) 159.
[58]        H. Bruschweiler, A.H. Yukagaku, Pure & Appl. Chem. 62 (1990) 781.
[59]        L. Zhang, Y. Liu, M.X. Xie, Y.M. Qiu, J. Chromatogr. A 1074 (2005) 1.
[60]        H.A. Oskooie, M.M. Heravi, M.H. Tehrani, F.K. Behbahani, O.M. Heravi, Phosphorus Sulfur Silicon Relat. Elem. 184 (2009) 1729.
[61]        Q.H. Zou, Y. Liu, M.X. Xie, J. Han, L. Zhang, Anal. Chim. Acta 551 (2005) 184.
该帖子作者被版主 天瑞俱乐部4积分, 2经验,加分理由:积极参加活动
天瑞俱乐部
feitian26
原文由 天瑞俱乐部(liushuitonghua) 发表:
感谢参赛
格式可以么?是否还需要修改。
天瑞俱乐部
社区=冬季=
lxx5052
委员长
这个研究成果的应用价值很高哦,经常在电视看到国外这爆炸那里爆炸。
titi
feitian26