浏览0 回复4 电梯直达
Insm_d2f9c757
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
各位大神,我想咨询一下用ICP-OES做样时,有的数据是红色的,而且在红色的数据后面有u或者x还有Z的字样,这些字母代表这个什么意思?谢谢
推荐答案:砂锅粥回复于2020/08/20
这个其实在仪器帮助有说明的,楼主不妨看看软件的帮助文件。

o Replicate signal overrange
QC--- QC test failed
S--- Standard addition failure
e--- Edited result
C--- Fitted background error seen on replicates, due to non-convergence
D ---Fitted background error seen on replicates, unable to be modelled due to insufficient spectral data
uncal Uncalibrated results
i--- No internal standard correction applied
n--- Interference correction not applied (line has a related interferent that was not successfully calculated)
a ----Solution overrange and automatically diluted
u--- Under-range (< blank)
v--- Under valid range (< minimum concentration)
x---- Extended calibration range (> maximum concentration)
b---- Line has a related interferent that is above the extended calibration range (IEC correction still applied)
为您推荐
专属顾问快速对接
获取验证码
立即提交
Insm_d2f9c757
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 砂锅粥(czcht) 发表:这个其实在仪器帮助有说明的,楼主不妨看看软件的帮助文件。o Replicate signal overrangeQC--- QC test failedS--- Standard addition failuree--- Edited resultC--- Fitted background error seen on replicates, due to non-convergenceD ---Fitted background error seen on replicates, unable to be modelled due to insufficient spectral datauncal Uncalibrated resultsi--- No internal standard correction appliedn--- Interference correction not applied (line has a related interferent that was not successfully calculated)a ----Solution overrange and automatically diluted u--- Under-range (v--- Under valid range (x---- Extended calibration range (> maximum concentration)b---- Line has a related interferent that is above the extended calibration range (IEC correction still applied)
好的,谢谢
砂锅粥
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
这个其实在仪器帮助有说明的,楼主不妨看看软件的帮助文件。

o Replicate signal overrange
QC--- QC test failed
S--- Standard addition failure
e--- Edited result
C--- Fitted background error seen on replicates, due to non-convergence
D ---Fitted background error seen on replicates, unable to be modelled due to insufficient spectral data
uncal Uncalibrated results
i--- No internal standard correction applied
n--- Interference correction not applied (line has a related interferent that was not successfully calculated)
a ----Solution overrange and automatically diluted
u--- Under-range (< blank)
v--- Under valid range (< minimum concentration)
x---- Extended calibration range (> maximum concentration)
b---- Line has a related interferent that is above the extended calibration range (IEC correction still applied)