METAL INTERFERENCES AND THEIR REMOVAL PRIOR TO THE DETERMINATION OF As(T) AND As(III) IN ACID
MINE WATERS BY HYDRIDE GENERATION ATOMIC ABSORPTION SPECTROMETRY
By R. Blaine McCleskey, D. Kirk Nordstrom, and James W. Ball
Water-Resources Investigations Report 03-4117
U.S. Department of the Interior
U.S. Geological Survey
Boulder, Colorado
2003
CONTENTS
Abstract.......................................................................................................................................................2
Introduction.................................................................................................................................................2
Methods of investigation.............................................................................................................................3
Sample collection and preservation.....................................................................................................3
Reagents .............................................................................................................................................3
Analytical apparatus ............................................................................................................................3
Analytical procedures.........................................................................................................................3
Determination of arsenic(T) .........................................................................................................4
Determination of arsenic(III)........................................................................................................4
Removal of interfering metals prior to arsenic(III) determination ...............................................4
Results and discussion ................................................................................................................................5
Selective reduction of arsenic(III).......................................................................................................5
Potential metal interferences on the determination of arsenic(T) .......................................................5
Potential metal interferences on the determination of arsenic(III)......................................................6
Cation exchange separation of iron(III) and copper............................................................................8
Accuracy and time stability of arsenic redox species in acid mine waters .........................................8
Accuracy of arsenic(T) determinations ........................................................................................8
Accuracy of arsenic(III) determinations.....................................................................................11
Time stability of arsenic redox species.......................................................................................11
Summary ...................................................................................................................................................12
References cited ........................................................................................................................................12
FIGURES
1-13. Graphs showing
1. Effect of NaBH4 concentration on the calibration curve in solutions containing only As(III)........5
2. Effect of NaBH4 concentration on the reduction of As(V) to As(III) in solutions containing
10 μg/L As(III) and varying As(V) concentrations .........................................................................5
3. As(T) determinations in the presence of potential interfering metals..............................................6
4. As(T) determinations in the presence of hydride-forming metals ...................................................6
5. As(III) determinations in the presence of potential interfering metals ............................................7
6. As(III) determinations in the presence of interfering metals ...........................................................7
7. As(III) determinations in the presence of hydride-forming metals .................................................8
8. Additive interference in the determination of As(III) in a solution containing 10 micrograms
per liter As(III) and 5 milligrams per liter Cu(II) ............................................................................8
9. Spike recovery determinations of As(T) using hydride generation atomic absorption
spectrometry....................................................................................................................................9
10. As(T) determinations by inductively coupled plasma-optical emission spectrometry and
hydride generation atomic absorption spectrometry for 20 samples collected from
Summitville Mine, Colo., 6 samples collected from Richmond Mine, Calif., and 3 samples
collected from the Penn Mine, Calif. ...............................................................................................9
11. Spike recovery determinations using hydride generation atomic absorption spectrometry for
As(III) in samples separated by cation exchange and not separated .............................................11
12. As(III) determinations using hydride generation atomic absorption spectrometry for samples
collected from the Summitville Mine ............................................................................................11
13. Time stability of As(III) for 45 surface and ground water samples from Yellowstone National
Park, Wyo.; Questa Mine site, N. Mex.; Summitville Mine site, Colo.; Richmond Mine,
Calif.; Penn Mine, Calif.; Ester Dome, Alaska; Fallon, Nev.; Mojave Desert, Calif.; and
Kamchatka, Russia reanalyze 3 to 15 months after the initial determination. The curved line
is a Gaussian Fit .............................................................................................................................12
TABLES
1. Cation exchange separation of Cu and Fe(III)..................................................................................9
2. Composition of acid mine water samples collected from Summitville Mine, Colo.;
Richmond Mine, Calif.; and Penn Mine, Calif. ............................................................................10
3. As(T) determinations in U.S. Geological Survey standard reference water samples.....................10
您可能想找: 原子吸收光谱(AAS) 询底价
tzl75
Explanation of abbreviations
--- (not analyzed or not measured)
°C (degrees Celsius)
AAS (atomic absorption spectrometry)
As(III) (arsenic(III) or arsenite)
As(V) (arsenic(V) or arsenate)
As(T) (total arsenic (As(III) plus As(V))
cm (centimeter)
conc. (concentration)
EDL (electrodeless discharge lamp)
FIAS (flow injection analysis system)
g (gas)
HCl (hydrochloric acid)
HGAAS (hydride generation-atomic
HNO3 (nitric acid)
KI (potassium iodide)
ICP-OES (inductively coupled plasma-optical
emission spectrometry)
ISE (ion-selective electrode)
μg/L (micrograms per liter)
M (moles per liter)
meq/L (milliequivalents per liter)
meq/mL (milliequivalents per milliliter)
mg/L (milligrams per liter)
min. (minute)
mL (milliliters)
mv (millivolt)
MPV (most probable value)
μL (microliter)
μm (micrometer)
μS/cm (microsiemens per centimeter at 25
degrees Celsius)
n (number of analyses)
NaBH4 (sodium borohydride)
nm (nanometer)
RSD (relative standard deviation)
s (standard deviation)
SRWS (standard reference water sample)
v/v (volume per volume)
w/v (weight per volume)
w/w (weight per weight)
tzl75
ABSTRACT
Hydride generation atomic absorption
spectrometry (HGAAS) is a sensitive and
selective method for the determination of total
arsenic (arsenic(III) plus arsenic(V)) and
arsenic(III); however, it is subject to metal
interferences for acid mine waters. Sodium
borohydride is used to produce arsine gas, but
high metal concentrations can suppress arsine
production.
This report investigates interferences of
sixteen metal species including aluminum,
antimony(III), antimony(V), cadmium,
chromium(III), chromium(IV), cobalt,
copper(II), iron(III), iron(II), lead,
manganese, nickel, selenium(IV),
selenium(VI), and zinc ranging in
concentration from 0 to 1,000 milligrams per
liter and offers a method for removing
interfering metal cations with cation exchange
resin. The degree of interference for each
metal without cation-exchange on the
determination of total arsenic and arsenic(III)
was evaluated by spiking synthetic samples
containing arsenic(III) and arsenic(V) with
the potential interfering metal. Total arsenic
recoveries ranged from 92 to 102 percent for
all metals tested except antimony(III) and
antimony(V) which suppressed arsine
formation when the antimony(III)/total
arsenic molar ratio exceeded 4 or the
antimony(V)/total arsenic molar ratio
exceeded 2. Arsenic(III) recoveries for
samples spiked with aluminum,
chromium(III), cobalt, iron(II), lead,
manganese, nickel, selenium(VI), and zinc
ranged from 84 to 107 percent over the entire
concentration range tested. Low arsenic(III)
recoveries occurred when the molar ratios of
metals to arsenic(III) were copper greater than
120, iron(III) greater than 70, chromium(VI)
greater than 2, cadmium greater than 800,
antimony(III) greater than 3, antimony(V)
greater than 12, or selenium(IV) greater than
1. Low recoveries result when interfering
metals compete for available sodium
borohydride, causing incomplete arsine
production, or when the interfering metal
oxidizes arsenic(III).
Separation of interfering metal cations
using cation-exchange prior to hydridegeneration
permits accurate arsenic(III)
determinations in acid mine waters containing
high concentrations of interfering metals.
Stabilization of the arsenic redox species for
as many as 15 months is demonstrated for
samples that have been properly filtered and
acidified with HCl in the field. The detection
limits for the method described in this report
are 0.1 micrograms per liter for total arsenic
and 0.8 micrograms per liter for arsenic(III).
tzl75
INTRODUCTION
Accurate determination of the redox
state of dissolved As is important for
interpreting its toxicity and mobility in the
environment. Dissolved As can form aqueous
species in several oxidation states, but in
natural waters occurs dominantly as the
inorganic oxyanions As(III) and As(V).
Several analytical methods use hydride
generation to selectively measure As redox
species, but high metal concentrations can
interfere with the determination of As(T)
(As(III) plus As(V)) and As(III) (Smith, 1975;
Welz and Melcher, 1984; Welsch and others,
1990; Creed and others 1996; Hageman and
Welsch, 1996). Accurate determination of
arsenic redox species in acid mine waters is
challenging because of elevated metal
concentrations frequently found in this type of
water (Nordstrom and Alpers, 1999). Sixteen
metal species including Al, Cd, Co, Cr(III),
Cr(VI), Cu(II), Fe(II), Fe(III), Mn, Ni, Pb,
Sb(III), Sb(V), Se(IV), Se(VI), and Zn were
evaluated as possible interferents on the
determination of As(T) and As(III). Because
As(III) and As(V) exist as oxyanions in
solution, interfering metal cations can be
removed from solution using cation-exchange
while maintaining the existing As(III) and
As(T) concentrations. Pre-analysis separation
of Fe(III) using cation-exchange resin for the
determination of As redox species by ion
chromatography in iron sulfate-sulfuric acid
media has been successfully performed (Tan
and Dutrizac, 1985). Cyanide also has been
used as a complexing agent to eliminate
interferences by metals in the determination
of As using hydride-generation (Jamoussi and
others, 1996). Interferences usually associated
with atomic absorption analysis are negligible
because arsine (AsH3(g)) is separated from
the sample matrix.
Results of accurate and precise
analytical methods can be invalidated by
deteriorated samples. Proper filtration and
preservation are critical for maintaining the
existing As(III)/As(T) ratio prior to analyses.
Wing and others (1987) observed that
filtration (0.1-μm or 0.45-μm), acidification
with HCl, and storage in the dark at 4°C
effectively preserved the As redox species for
up to 8 months in ground waters collected
near Fallon, Nev. Except for Fe(III) and
hydrogen sulfide (H2S), laboratory studies
demonstrate that rates of oxidation of As(III)
and reduction of As(V) in aqueous solution
with redox agents common to natural waters
are slow (Cherry and others, 1979). Filtering
the sample removes colloidal material and
bacteria that can affect the dissolved
As(III)/As(T) ratio (Gihring and others, 2001;
Wilkie and Hering, 1998). Acidification
prevents precipitation of Fe hydroxides that
can coprecipitate or adsorb As (Gao and
others, 1988; Wilkie and Hering, 1996).
Excluding light inhibits potential
photochemical reactions (Emett and Khoe,
2001; Hug and others, 2001) and storing the
samples at 4°C slows chemical reactions.
tzl75
To address the need for more accurate
determinations of As(T) and As(III) by
spectrometry (HGAAS), the U.S. Geological
Survey (USGS) evaluated metal interferences
and their removal in acid mine waters. This
report (1) describes sample collection,
preservation, and analytical procedures for the
accurate determination of As(T) and As(III)
using HGAAS; (2) identifies metals and the
concentration at which each interferes in the
determination of As(T) and As(III); (3)
demonstrates that Cu(II) and Fe(III), both of
which interfere in the determination of As(III)
by HGAAS, can be removed from solutions
using cation-exchange while maintaining
existing As(III)/As(T) ratios; (4) applies the
method to acid mine waters collected from the
Summitville Mine, Colo.; Penn Mine, Calif.;
and Richmond Mine, Calif.; and (5) presents
As redox species time stability data for 45
surface- and ground-water samples collected
during 1999-2002 from Yellowstone National
Mine, Calif.; Penn Mine, Calif.; Ester Dome,
and Kamchatka, Russia.
tzl75
METHODS OF INVESTIGATION
Sample Collection and Preservation
Water samples collected in the field were
pumped from the source through a 0.1 μm
tortuous-path filter, acidified to pH less than 2
with hydrochloric acid (typically 2 mL 6-M
HCl per 250 mL sample), and stored in acidwashed
opaque bottles at 4°C (To and others,
1999).
Reagents
All reagents were of purity at least equal
to the reagent-grade standards of the
American Chemical Society. Double-distilled
water and re-distilled or trace metal grade
acids were used in all preparations. The
following reagents were used for the As(T)
and As(III) HGAAS procedure: 10 percent
(w/v) KI from Aldrich; 10 percent (w/v) LAscorbic
Acid from Aldrich; NaOH from
Fisher; NaBH4 from Fisher; trace metal grade
HCl from Fisher, Assay w/w 35 to 38 percent;
1,000 mg/L As(III) from High Purity
Standards; 1,000 mg/L As(V) from High
Purity Standards; and cation-exchange resin,
AG 50W-X8, 20-50 mesh, H+ form from Bio-
Rad.
The following single-element standards
(1 to 10,000 mg/L) were prepared for
interference studies: CuSO4⋅5H2O was used to
prepare Cu(II); FeSO4⋅7H2O was used to
prepare Fe(II); Fe2(SO4)3 was used to prepare
Fe(III); Zn(NO3)2⋅6H2O was used to prepare
Zn(II); MnSO4⋅H2O was used to prepare
Mn(II); Al(NO3)3⋅9H2O was used to prepare
Al(III); Ni(SO4)⋅6H2O was used to prepare
Ni(II); CrCl3⋅6H2O was used to prepare
Cr(III); K2Cr2O7 was used to prepare Cr(VI);
CoCl2⋅6H2O was used to prepare Co(II); Cd
powder was used to prepare Cd(II); NaSeO4
was used to prepare Se(VI); Sb2O3 was used
to prepare Sb(III); SbCl5 was used to prepare
Sb(V); and Pb metal was used to prepare Pb.
The Se(IV) standard was prepared by adding
an equal volume of concentrated HCl to a
portion of Se(VI) standard and heating at
90°C for 20 minutes.
tzl75
Analytical Apparatus
An atomic absorption spectrophotometer
(Perkin-Elmer (PE) – AAnalyst 300) with an
electrically heated quartz cell having a path
length of 15-cm inline with a flow injection
analysis system (FIAS; PE – FIAS 100), an
autosampler (PE – AS90), and an arsenic
electrodeless discharge lamp (EDL) attached
to an EDL power supply (PE – EDL System
2) were used. The following spectrometer
parameters were used: EDL current: 380 mv;
wavelength: 193.7 nm; slit: 0.7 nm. Peak
height was used for data processing. The
following FIAS parameters were used: carrier
gas: Ar; cell temperature: 900°C; sample
loop: 500 μL; carrier solution: 10 percent
(v/v) HCl; reducing agent: 0.25 percent
(As(T)) or 0.03 percent (As(III)) NaBH4 in
0.05 percent NaOH; carrier solution flow rate:
10 mL/min; reductant flow rate: 5 mL/min.
Sodium borohydride was prepared daily and
filtered through a 0.45 μm polyvinylidene
fluoride filter membrane using a vacuum
pump.
4077
嗯,不错。
现在国内做得也很好,不仅可以测定总量,三价,五价,而且有机砷也可以测定。
tzl75
Analytical Procedures
Hydride formation occurs when As(III)
in an acidic solution reacts with NaBH4
according to the following reaction:
3NaBH4 + 4H3AsO3 →
4AsH3(g) + 3H3BO3 + 3NaOH (1)
For the determination of As(T), As(V) is prereduced
to As(III) using KI and L-ascorbic
acid.
All water samples and standards were
prepared in 25-mL volumetric flasks. The
method detection limits of the HGAAS
analytical procedure used at the USGS
National Research Laboratory in Boulder,
Colo., are 0.1 μg/L for As(T) and 0.8 μg/L for
As(III).
Determination of Arsenic(T)
1. A solution containing 10 percent (w/v)
KI and 10 percent (w/v) L-ascorbic
acid was prepared.
2. To each 25-mL volumetric flask, 5 mL
concentrated HCl and 2.5 mL of the
KI−L-ascorbic acid solution prepared
from step 1 were added.
3. The As(III) standard stock solution
was used to prepare standards ranging
in concentration from 0 to 10 μg/L As.
4. To demonstrate completeness of
reduction, synthetic samples
containing varying As(III)/As(V)
ratios were prepared.
5. A volume of sample (up to 16.5 mL)
that will yield 1 to 10 μg/L As(T)
when diluted to volume was
transferred to a 25-mL volumetric
flask.
6. Appropriate spike-recovery samples
were prepared by adding both As(III)
and As(V) to selected samples to
demonstrate reduction of As(V) and
AsH3(g) production.
7. The HGAAS analyses were performed
a minimum of 30 minutes after sample
preparation.
tzl75
Determination of Arsenic(III)
1. Twenty-five mL of concentrated HCl
was added to each 25-mL volumetric
flask.
2. As(III) standards ranging in
concentration from 0 to 20 μg/L were
prepared.
3. To demonstrate that As(V) was not
reduced, synthetic samples were
prepared containing several
As(III)/As(V) ratios.
4. A volume of sample (up to 24.75 mL)
that will yield 2 to 20 μg/L As(III)
flask.
5. Appropriate spike-recovery samples
and As(V) to demonstrate that As(V)
is not reduced and that adequate
AsH3(g) production occurs.
6. The HGAAS analysis was performed
immediately.
Removal of Interfering Metals Prior
To Arsenic(III) Determination
For samples having a Cu(II)/As(III)
molar ratio greater than 120, Fe(III)/As(III)
molar ratio greater than 70, Cd/As(III) molar
ratio greater than 800, or for samples with low
As(III) spike recoveries, the sample was
mixed with cation-exchange resin (AG 50WX8,
20-50 mesh, H+ form) to remove
interfering metals prior to determination of
As(III). The following procedure was used to
remove interfering metals and regenerate the
resin:
1. Approximately 100 grams dry resin
was transferred to a 250-mL
polyethylene bottle. The resin was
washed 2 times with 100 mL 2-M HCl
followed by 5 washes with distilled
water. The excess water was decanted.
2. The resin was dried in an oven at
100°C.
3. To a 50-mL polyethylene bottle, 12.5
grams dried resin and 25 mL sample
were added. The mixture was
periodically shaken for 30 minutes.
4. The sample was separated from the
resin by decanting it into an acidwashed
15-mL polyethylene bottle.
This sample was used to determine the
As(III) concentration.
tzl75
RESULTS AND DISCUSSION
Selective Reduction of Arsenic(III)
During the As(III) determination, As(V)
can be reduced to AsH3(g) by the NaBH4
reductant causing an overestimation of As(III)
concentration. At a constant carrier
concentration and FIAS flow rate, increasing
the NaBH4 concentration increases the
sensitivity of the As(III) determination (fig.
1); however, when the NaBH4 concentration
exceeds 0.0625 percent, As(V) is reduced,
concentration (fig. 2). To ensure selective
reduction of As(III), the NaBH4 concentration
is maintained at 0.03 percent for the As(III)
determination for analyses described in this
report.
Potential Metal Interferences on the
Sixteen metal species including Al, Cd,
Co, Cr(III), Cr(VI), Cu(II), Fe(III), Fe(II),
Mn, Ni, Pb, Sb(III), Sb(V), Se(IV), Se(VI),
and Zn were evaluated as possible interferents
on the determination of As(T). The
interference of each metal was evaluated by
spiking synthetic samples containing 6 μg/L
As(III) and 6 μg/L As(V) with 0.002 to 1,000
mg/L of the potential interfering metal and
then determining the As(T) concentration
using the HGAAS procedure (without cation
exchange treatment). Arsenic(T) recoveries
ranged from 92 to 102 percent for individual
solutions spiked with Al, Cd, Co, Cr(VI),
Cr(III), Cu(II), Fe(II), Fe(III), Mn, Ni, or Zn
(fig. 3). For solutions spiked with a
combination of metals including Al, Cd, Co,
Cr(III), Cr(VI), Cu(II), Fe(II), Fe(III), Mn, Ni,
and Zn, each up to a concentration of 500
mg/L, As(T) recoveries ranged from 97 to
102 percent. Dissolved Cu(II) was reduced to
Cu metal during the pre-reduction step. The
precipitated metal is visible at Cu
concentrations greater than 100 mg/L and the
liquid phase was decanted and analyzed with
little change in As(T) concentration.
Antimony(III) and Sb(V) were the only
hydride-forming species found to interfere
with the formation of AsH3(g) (fig. 4).
Antimony(III) and Sb(V) substantially
suppressed AsH3(g) formation by reacting
with the NaBH4 when the Sb(III)/As(T) molar
ratio exceeded 4 or the Sb(V)/As(T) molar
ratio exceeded 2. In most waters,
concentrations of Sb(III) and Sb(V) are much
lower than As(T) concentrations and are
likely to be diluted (As(T) less than 10 μg/L)
to a concentration that will not interfere with
the As(T) determination. Alkali and alkaline
earth elements do not interfere with the As(T)
determination (Smith, 1975).