One metal ball which surface has paint,it is immersed in lime slurry and the paint will gradually solve into the lime slurry. I want to identify the components of the paint in the lime slurry. So i should compare between the lime slurry without paint and the lime slurry with paint erosion.

I prepare the two samples for gc-MS test. First I adjust the pH value by adding sulfic acid to 6.0-7.0.  Then I extract them with Hexane solvent and concentrate the extraction.      The two sample be injected to gc. The amount is about 1-2(ul) micro liter.The heating program is following: ramp rate varies from 5C/min to 50C/min; target temperature 300 or 320C. I repeat many times(about 100 times). every time i got the different peak table even all parameters are the same.From the Mass spectrum picture, I found many compounds contained  m/z 73,147,221,281,327, 404, 454. I think it may be due to DB-5 phase flow. How I should do if you make sure the DB-5 phase flow?
If the peak table has many compounds, but the similarities of compounds are very low(less than 750) for most of compunds in the peak table. The name of  many compounds shows UNKNOW. What is problem with our gc/MS?

专家答疑:
It sounds like the column in the gc is "bleeding."  You should proably check the column for excessive bleed by just ramping the column without injecting a sample.  This will give you a good indication of the state of the column.  After that, you can just inject a microliter or two of the hexane extraction solvent and compare that chromatogram
with the one from above.

If there are significant differences (i.e. more peaks in the solvent injectin than in the blank gc ramp) then you mught have "junk" in your liner in the gc inlet.  These should be changed ever so often, depending on how many samples you
run, their cleanliness, etc.  Some labs replace them every day, others after so man injections, say, 50 to 100.

As for the differences in the peak tables, it might be that you are not holding your final oven temperature long enough to allow the compounds to elute from the
column.  You should hold it at the upper temperature until you do not see any more peaks.
为您推荐
您可能想找: 气质联用(GC-MS) 询底价
专属顾问快速对接
立即提交
icefirel
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
easyboy
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
nikang3148
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
楼主在美国工作,接触的人都讲英语,大家借此机会锻炼英语吧
nikang3148
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
专家好像是在说可能是柱子流失,并建议楼主进行延长温度保持时间使得物质尽可能从柱子流出吧
猜你喜欢最新推荐热门推荐更多推荐
品牌合作伙伴