Dediacted monochromator-based UV-Vis spectrometers, such as the Nicolet Evolution series, are widely used for UV-Visible spectroscopic measurements. Research -grade Nicolet FT-IR spectrometers can be optionally configured to cover the near-UV and nearly full range of visible spectroscopic regions (27000-128000cm-1). This configuration contains a quartz-halogen light source (27000-2000cm-1), a quartz beamsplitter (27,000-2,800cm-1), a silicon detector (2,700-8,600cm-1), and wide and narrow blue filters. Blue filters are used to optimize thje spectral result, to block any unwanted energy from reaching the detector (such as interference from the He-Ne laser at 15798cm-1 or energy at lower wavenumbers), and to prevent folding of adjacent spectral regions into the desired region. Single-beam spectra of the quartz-halogen source from 28,000 to 2,000cm-1 with different combinations of detector (Si or PbSe), beamsplitter (Quartz or CaF2) and optical filter (narrow blue, wide blue or red) are shown in Figure 2.6. The diffuse reflectance spectrum of yttrium chloride from 25,000 to 2,000cm-1 is shown in Figure 2.7.10 To optimize the near-IR regions, a red filter was used for spectral region 14,000-9,000cm-1, a CaF2 beamsplitter, and a PbSe detector were used for spectral region 9,000-2,000cm-1.