3. FUME HOOD (通风柜)
Bench-type fume hood is the main type in CCIQ laboratory.
台式通风柜是CCIQ实验室主要的通风柜。
Fume hood face velocity shall be 100fpm (0.5m/s), actual exhaust volume shall be 900 m3/h to 1150 m3/h according to SEFA & AHRAE recommendations. Canopy hood shall have 600m3/h each outlet to ensure weak negative exhaust. The volume of other exhaust outlet for vented cabinet, vacuum oven etc. should comply with equipment requirement.
通风柜表面风速为0.5m/s,根据SEFA和AHRAE的要求,通风柜排气量在900 m3/h 到 1150 m3/h之间。每个排气罩排气量在600m3/h。另外,在特殊存储柜和仪器上方排气罩的排气量满足设定的要求。每个排气口保持负压。