781  OPTICAL ROTATION 旋光


Many pharmaceutical substances are optically active in the sense that they rotate an incident plane of polarized light so that the transmitted light emerges at a measurable angle to the plane of the incident light. This property is characteristic of some crystals and of many pharmaceutical liquids or solutions of solids. Where the property is possessed by a liquid or by a solute in solution, it is generally the result of the presence of one or more asymmetric centers, usually a carbon atom with four different substituents. The number of optical isomers is 2n, where n is the number of asymmetric centers. Polarimetry, the measurement of optical rotation, of a pharmaceutical article may be the only convenient means for distinguishing optically active isomers from each other and thus is an important criterion of identity and purity.
很多药用物质具有光学活性,他们能够使入射的平面偏振光旋转,这样射出的光线与入射光线的角度形成可测量的角度。这个性质是一些结晶状物体或者药用液体或固体的溶液所具有的特性。在某种液体或者溶液中的溶质具有该性质的情况下,这通常是存在着一个或者多个不对称中心的结果,经常是一个碳原子与四个不同的取代基。光学异构体的数量是2 n,在此n是不对称中心的数量。药用物质的旋光测定法,即测定光学旋转的方法,也许是用于区分具有光学活性的异构体的方法中唯一简便的方法。因此,该方法是鉴别和纯度的一个重要标准。

Substances that may show optical rotatory power are chiral. Those that rotate light in a clockwise direction as viewed towards the light source are dextrorotatory, or (+) optical isomers. Those that rotate light in the opposite direction are called levorotatory or (–) optical isomers. (The symbols d- and l-, formerly used to indicate dextro- and levorotatory isomers, are no longer sanctioned owing to confusion with D- and L-, which refer to configuration relative to D-glyceraldehyde. The symbols R and S and  and  are also used to indicate configuration, the arrangement of atoms or groups of atoms in space.)
能够表现出旋光能力的物质是手性的。那些朝向光源观察时按顺时针方向旋转光线的物质是右旋性的或者(+)光学异构体。(其符号d- 和 l-,原先是用来表示右旋的和左旋的异构体,现在由于容易与D-和L-混淆而不再使用,这两个符号涉及与D-甘油醛相关的结构)。符号R和S以及 和 也都是用于表示结构、原子或原子组在空间中的排列。

The physicochemical properties of nonsuperimposable chiral substances rotating plane polarized light in opposite directions to the same extent, enantiomers, are identical, except for this property and in their reactions with other chiral substances. Enantiomers often exhibit profound differences in pharmacology and toxicology, owing to the fact that biological receptors and enzymes themselves are chiral. Many articles from natural sources, such as amino acids, proteins, alkaloids, antibiotics, glycosides, and sugars, exist as chiral compounds. Synthesis of such compounds from nonchiral materials results in equal numbers of the enantiomers, racemates. Racemates have a net null optical rotation, and their physical properties may differ from those of the component enantiomers. Use of stereoselective or stereospecific synthetic methods or separation of racemic mixtures can be used to obtain individual optical isomers.
具有在相反方向旋转平面偏振光到同样程度的理化特性的不可重叠手性物质,即对映结构体,除了上述特性以及与其他手性物质进行的反应的区别之外,是完全相同的。由于生物受体和酶自身均是手性的,对映异构体经常在药理与毒理方面展现出深刻的差异。许多天然来源的物品,例如氨基酸、蛋白质、生物碱、抗生素、糖苷、蔗糖等,以手性化合物的形式存在。使用非手性物料合成此类化合物导致等量的对映异构体和外消旋物。外消旋物有净无效旋光(net null optical rotation),并且他们的物理特性可能与那些组成部分中的对应结构体不同。可以使用立体选择或立体专一的合成方法,或者外消旋混合物的分离,来获得单独的光学异构体。

Measurement of optical rotation is performed using a polarimeter.* The general equation used in polarimetry is:
旋光的测量使用一个旋光计来完成。旋光测定法中通用的公式为:

where [ ] is the specific rotation at wavelength  , t is the temperature, a is the observed rotation in degrees ( ), l is the pathlength in decimeters, and c is the concentration of the analyte in g per 100 mL. Thus, [ ] is 100 times the measured value, in degrees ( ), for a solution containing 1 g in 100 mL, measured in a cell having a pathlength of 1.0 decimeter under defined conditions of incident wavelength of light and temperature. For some Pharmacopeial articles, especially liquids such as essential oils, the optical rotation requirement is expressed in terms of the observed rotation, a, measured under conditions defined in the monograph.
其中,[ ]是在波长 条件下的比旋度,t是温度,a是观测到的以度( )为单位的旋转,l是以分米为单位的通道长度,c是以克/100mL为单位被分析物的浓度。因此,当100mL中含有1克的溶液,在通道长度为1.0分米的单位中,并使用规定的入射光波长和温度的情况下,[ ]是测量值的100倍。对于某些药典物质来说,尤其是香精油之类的液体,旋光的规定表达为在该各论中所规定的条件下进行测定,所观察到的旋转,a。

Historically, polarimetry was performed using an instrument where the extent of optical rotation is estimated by visual matching of the intensity of split fields. For this reason, the D-line of the sodium lamp at the visible wavelength of 589 nm was most often employed. Specific rotation determined at the D-line is expressed by the symbol:
历史上,旋光测定法检测所使用的仪器中,旋光的程度是通过肉眼观察分离区域的亮度。由于这个原因,可见波长589nm的纳灯的D-线是最经常使用的。用D-线检测的比旋度的表示方法为:

and much of the data available are expressed in this form. Use of lower wavelengths, such as those available with the mercury lamp lines isolated by means of filters of maximum transmittance at approximately 578, 546, 436, 405, and 365 nm in a photoelectric polarimeter, has been found to provide advantages in sensitivity with a consequent reduction in the concentration of the test compound. In general, the observed optical rotation at 436 nm is about double and at 365 nm about three times that at 589 nm. Reduction in the concentration of the solute required for measurement may sometimes be accomplished by conversion of the substance under test to one that has a significantly higher optical rotation. Optical rotation is also affected by the solvent used for the measurement, and this is always specified.
并且,很多得到的数据用次形式表示。使用较低的波长,例如那些利用光电旋光计过滤掉汞灯线在578、546、436、405、365nm的最大能见度所产生的,已经被发现有利于增加检验敏感性,并最终导致供试品浓度的降低。通常情况下,观察到的旋光在436nm条件下大约是两倍于589nm,而在365nm条件下是三倍。检验必须的溶质浓度的减少有时候可以通过将供试品转换为具有显著高旋光性的物质来实现。旋光还受检验所用溶剂的影响,而此溶剂总是作出明确规定的。


USP30-NF25比旋度《781》
该帖子作者被版主 happyjyl10积分, 2经验,加分理由:感谢分享
为您推荐
您可能想找: 气相色谱仪(GC) 询底价
专属顾问快速对接
立即提交
可能感兴趣
暴走追风
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
It is now common practice to use other light sources, such as xenon or tungsten halogen, with appropriate filters, because these may offer advantages of cost, long life, and broad wavelength emission range, over traditional light sources.
现在,使用其他光源,例如氙或者钨卤素,并使用适当的过滤器已经成为常规操作,因为这些光源与传统光源相比,可以提供在费用、使用寿命、宽波长发射范围等方面的优势。

Specific Rotation— The reference Specific rotation  781S  in a monograph signifies that specific rotation is to be calculated from observed optical rotations in the Test solution obtained as directed therein. Unless otherwise directed, measurements of optical rotation are made at 589 nm at 25 . Where a photoelectric polarimeter is used, a single measurement, corrected for the solvent blank, is made. Where a visual polarimeter is employed, the average of no fewer than five determinations, corrected for the reading of the same tube with a solvent blank, is used. Temperature, which applies to the solution or the liquid under test, should be maintained within 0.5  of the stated value. Use the same cell for sample and blank. Maintain the same angular orientation of the cell in each reading. Place the cell so that the light passes through it in the same direction each time. Unless otherwise specified, specific rotation is calculated on the dried basis where Loss on drying is specified in the monograph or on the anhydrous basis where Water is specified.
比旋度:在各论中提到“比旋度<781S>”意味着比旋度的计算将使用该各论中指定的供试液所观察到的旋光度。除非另外指出,旋光度的测量在25 、589nm的条件下进行。如果使用光电旋光计,只需一次测量并以溶剂作空白校正即可。如果旋光计是肉眼观察的,需要至少5次检测取平均值,并以溶剂作校正空白。供试溶液或者液体的温度应该保持在要求数值加减0.5 的范围内。进行样品检测和空白试验时,需使用同一个单元。在每一次读取数据的时候,使该单元保持同样的的成角方位(angular orientation)。放好单元,以便每次光线从同一个角度穿过它。除非另外指出,当该各论中要求干燥失重时,比旋度按干燥品计算,或者当该各论中要求水分时,在按无水计算。

Optical rotation of solutions should be determined within 30 minutes of preparation. In the case of substances known to undergo racemization or mutarotation, care should be taken to standardize the time between adding the solute to the Solvent and introduction of the solution into the polarimeter tube.
溶液的旋光度应该在溶液配制好30分钟内进行测定。对于检测已知具有外消旋或变旋作用的物质,应当小心地采取措施对向溶剂中加入溶质和将溶液加入旋光计管道的间隔时间进行标准化。

Angular Rotation— The reference Angular rotation  781A  in a monograph signifies, unless otherwise directed, that the optical rotation of the neat liquid is measured in a 1.0-dm tube at 589 nm at 25 , corrected for the reading of the dry empty tube.
成角旋光:当各论中提到“成角旋光<781A>”,除非有另外的规定,纯液体的旋光是在1.0分米管道、589nm波长、25 条件下检测的,并以干燥空管的读数进行校正。
________________________________________
*  Suitable calibrators are available from the Office of Standard Reference Materials, National Institute of Standards and Technology (NIST), Gaithersburg, MD 20899, as current lots of Standard Reference Materials, Dextrose and Sucrose. Alternatively, calibration may be checked using a Polarization Reference Standard, which consists of a plate of quartz mounted in a holder perpendicular to the light path. These standards are available, traceable to NIST, from Rudolph Research Analytical, 354 Route 206, Flanders, NJ 07836, or from Rudolph Instruments, Inc., 40 Pier Lane, Fairfield, NJ 07004-2113.
适合的校正设备可以从国立标准和技术研究院标准物质办公室获得。或者,校准可以通过使用一个偏振标准品来进行,该标准品由一片装在垂直于右侧通道的架子上的石英构成。这些标准品可以从以下2处得到,并可追索到国立标准和技术研究院。

Auxiliary Information— Staff Liaison : Antonio Hernandez-Cardoso, B.S., Scientist, Latin American Specialist
Expert Committee : (GC05) General Chapters 05
USP30–NF25 Page 314
Pharmacopeial Forum : Volume No. 27(4) Page 2819
Phone Number : 1-301-816-8308

USP30-NF25比旋度《781》
poorlittle
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 zhk521 发表:
These standards are available, traceable to NIST
这些标准品可以从以下2处得到,并可追索到国立标准和技术研究院。


I think that it should be追溯, not 追索

In metrology, “traceable to” is generally translated as 溯源至.
何当奇
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 poorlittle 发表:
原文由 zhk521 发表:
These standards are available, traceable to NIST
这些标准品可以从以下2处得到,并可追索到国立标准和技术研究院。