1. Wavelength of UV-Visible Detector (HPLC)— Deviations from the wavelengths specified in the method are not permitted.
2. The procedure specified by the detector manufacturer, or another validated procedure, is to be used to verify that error in the detector wavelength is, at most, ±3 nm.
1. Wavelength of UV-Visible Detector (HPLC)— Deviations from the wavelengths specified in the method are not permitted. 检测器设置的波长必须和方法上的一致 2. The procedure specified by the detector manufacturer, or another validated procedure, is to be used to verify that error in the detector wavelength is, at most, ±3 nm.