6.5.3 VLD Calibration and Polarization Modulation-Based Dichroic Difference
The VLD calibration measurements are performed by replacing the sample with a reference linear polarizer. The linear polarizer is oriented either parallel or perpendicular to the polarizer positioned in front of the PEM. Thus, the palizer represents a sample that completely transmits one polarization and totally absorbs the other. The respective calibration reference spectra R‖,‖, and R‖,⊥, derived from Equation (6-11) by suppressing T‖ or T⊥, are given by the following relations: