3.3.6 Step-Scan Phase Modulation (S2ΦM) FT-IR PAS: Experimental
PAS Reference. Since only the absorbed IR light generates thermal waves and contributes to PA detector signals, the PA single-beam spectrum is an uncorrected "absorption" spectrum of the sample. The variations over the spectral region in source intensity, beam splitter efficiency, and detector response must be corrected using a reference sample that absorbs strongly at all wavenumbers. A common practice in both continuous- and step-scan PAS depth profiling is to use a strongly absorbing glassy carbon (or a highly concentrated carbon black-filled rubber, minimum carbon content 60%) as a reference. Since the reference absorbs strongly at all wavenumbers in the mid-IR region, its magnitude spectrum can be used as a reference to normalize PA spectra, correcting all those variations mentioned above. In addition, since almost all absorption occurs at surface of the strong absorber, it can be used as a surface reference for calibrating the phase of a step-scan phase modulation depth profiling experiment. When this reference is in place, the relative ΦM phase setting button, as shown in Figure 3.6, is pressed to maximize PA signals in the in-phase (I) channel and subsequently the quadrature (Q) channel is minimized. After this phase angle calibration is done, the ΦM phase setting will be kept unchanged for running a sample under the same modulation frequency. The resulting in-phase spectrum of the sample will therefore enhance the surface absorption features and the quadrature spectrum will enhance those from the substrate or deeper part of the sample.