主题:【讨论】小肽的子离子比母离子还大?为什么?
浏览0 回复2 电梯直达
wqrwxm
有一篇文献用MRM 方法测定CYP3A4, 肽段是LSLGGLLQPEKPVVLK。文献上选择母离子 m/z=564.6, 子离子m/z=746。
为什么母离子比子离子还大?
难道是母离子电荷比子离子多?
原文链接:http://www.sciencedirect.com/science/article/pii/S0006295214005838
原文方法如下:
The CYP protein quantification assays were performed on a triple-quadrupole $$lc-MS instrument (Agilent 6460A) coupled to an Agilent 1290 Infinity $$lc system (Agilent Technologies), in ESI positive ionization mode. Approximately 2 μg of the trypsin digest (5 μl) was injected onto the column (Kinetex 1.7 μ, C18 100A; 100 × 2.1 mm, Phenomenex, Torrance, CA). Mobile phases consisted of water (A) and acetonitrile (B) (both containing 0.1% formic acid) at a flow rate of 0.3 ml/min. The gradient was 3% mobile phase B for 2.0 min, followed by gradient elution program with mobile phase B concentration of 3 to 18% (2.0-4.0 min), 18 to 22% (4.0-8.0 min), 22 to 24% (8.0-10 min), 24% (10-12.2 min), 24 to 34% (12.2-12.5 min) and 34 to 38% (12.5-15.0 min). This was followed by washing with 80% mobile phase B for 0.9 min and re-equilibration for 4.9 min. Surrogate peptides and internal standards were monitored in multiple reaction monitoring (MRM) mode using instrument parameters provided in Table 1. HP$$lc-MS/MS data were processed using the MassHunter (Agilent Technologies) and Skyline software.
Table 1.
Optimized MS/MS parameters used for quantification of surrogate peptides and internal standards of CYP enzymes in trypsin digested microsomes.
EnzymePeptidePrecursor ion (m/z)Product ion (m/z)Fragmentor (V)Collision energy (eV)LLOQ (fmol on-column)
CYP3A4LSLGGLLQPEKPVVLK564.674610090.3
  564.6689.410010
 LSLGGLLQPEKPVVLK567.37501009
  567.3693.41009
CYP3A5DSIDPYIYTPFGTGPR900.0995.7130281
  900.0684.913020
 DSIDPYIYTPFGTGPR904.91005.513028
  904.9689.913028
CYP3A7FNPLDPFVLSIK695.4918.4110160.1
  695.4564.711016
 FNPLDPFVLSIK699.4926.511016
  699.4568.811016
CYP2C9GIFPLAER451.9585.47570.15
  451.9366.8757
 GIFPLAER456.8595.3757
  456.8371.7757
CYP2C19IYGPVFTLYFGLER838.01145.4145281
  838.0998.314528
 IYGPVFTLYFGLER843.01155.614528
  843.01008.514528
CYP2A6GTGGANIDPTFFLSR777.1982.5130220.4
  777.1867.213026
 GTGGANIDPTFFLSR781.9992.513026
  781.9877.513026
CYP2D6GTTLITNLSSVLK673.9974.4110160.6
  673.9861.311016
 GTTLITNLSSVLK677.9982.411016
  677.9869.311016
CYP reductaseFAVFGLGNK476.9734.310090.05
  476.9635.31009
 FAVFGLGNK480.8742.41009
  480.8643.41009
Underlined residues represent the internal standards with the labeled [13C615N2]-lysine or [13C615N4]-arginine.
shmyd03