BEAM INJECTION FLAME FURNACE ATOMIC ABSORPTION SPECTROMETRY
A NEW FLAME AAS METHOD

The sensitivity of flame atomic absorption spectrometry
measurements can be improved significantly by
increasing the efficiency of aerosol generation/transport
and prolonging the residence time of the free
analyte atoms in the absorption volume. Both can be
achieved using the new Beam Injection Flame Furnace
AAS (BIFF-AAS). For the first time a continuous liquid
sample introduction into a flame-heated tube furnace
has become possible by sample transportation via a
very narrow liquid jet. With this technique it is possible
to introduce the sample over a distance of 10 cm into a
flame-heated metal or ceramic furnace (Fig. 1). The
liquid jet is created by means of an HPLC pump (50 to
400 bar), a gas pressure pump (2 to 10 bar), a
peristaltic pump (< 2 bar) or a novel 6 bar flow-system
which feeds a smooth jet nozzle with a diameter of 50
:m or smaller. The liquid jet passes a small sample
introduction hole (2 mm diameter) and is then
impacted on the opposite inner wall of the tube furnace
leading to an immediately vaporizing aerosol (jet
impact vaporization, JIV). The arrangement containing
a peristaltic pump for the beam generation is shown in
Fig. 2. The complete introduction of the whole sample
directly at the place of atomization and the extended
residence time inside the absorption volume (flame
furnace) result in a considerable improvement in the
power of detection compared with the standard
working mode of flame AAS (sample introduction via a
pneumatic nebulizer).
为您推荐
您可能想找: 原子吸收光谱(AAS) 询底价
专属顾问快速对接
立即提交
tzl75
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
The experiments showed that a beam is stable over a
distance of at least 15 cm if its speed is higher than
about 12 m/s. Using a nozzle of 50 :m inner diameter,
a 1.42 mL/min flow rate is necessary to reach this
speed. The typical shape of a smooth jet nozzle is a
cylinder with a ratio length/diameter > 3. The pressure drop of such a nozzle is less than 0.5 bar which
allows the use of low-pressure pumps, for instance a peristaltic pump, for the beam generation.
The requirements for the material of the tube furnace include good stability and chemical resistance
against the corrosive conditions of an acetylene/air flame and water and acid vapors at temperatures higher
than 1000 /C.
Suitable tube materials are pure nickel, super alloys or non-porous Al2O3 ceramic. To increase the
temperature inside the furnace and to achieve a more reducing atmosphere, a tube with 4 or more
additional holes located on the bottom side is used.
Currently, 25 elements have been determined
using BIFF-AAS, of which 17 elements (Ag, As,
Au, Bi, Cd, Cu, Hg, In, K, Pb, Pd, Rb, Sb, Se,
Te, Tl, Zn) exhibit an improved power of
detection from 6 fold to 202 fold compared to
conventional FAAS [1]. The standard deviation
amounts to 1.7 - 4.0 % (N = 12; 50 :L
samples). Fig. 3 shows the signals of Cd sample
in the low :g/L range measured with BIFFAAS.
tzl75
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵

The sample is injected via a sample introduction
valve into a carrier stream as in all flow injection
techniques including HPLC. Because of this,
even sample volumes as small as 10 :L can be
easily applied. Fig. 4 shows the determination of
copper with increasing sample volumes and
using BIFF-AAS as well as conventional flame-
AAS. With 10 :L samples a clear Cu signal
appears in case of BIFF-AAS but not when
using the normal flame AAS. Working with a
peristaltic pump allows a continuous sample
introduction to be maintained. Generating the
liquid jet with the aid of a high-pressure pump,
BIFF-AAS can be applied as a simple, efficient
interface between HPLC separation techniques
and elemental determinations by flame-AAS.

tzl75
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵


To fill the gap in the pressure range between
HPLC and peristaltic pumps a novel pump
system based on a 6 bar low cost diaphragm
pump has been developed (Fig. 5). The unavoidable
strong pulsation has been overcome by
using 50 m (2 mm i.d.) of highly flexible silicone
tubing as a pulse suppression coil. This results
in a smooth pulse free continuous flow whereby
the pump delivers 100 ml/min in circulation. For
analytical flow injection a restrictor is connected
to a splitting point in the cycle to tap off the
flow needed, down to 0.2-5.0 ml/min. The
higher pressure of the diaphragm pump
compared to the peristaltic pump permits the
use of common HPLC-columns or low-pressure
chromatography columns for online-sample
pretreatment like preconcentration of traces or
matrix separation. This system was successfully
applied for the beam generation in BIFF-AAS
and for on-line sample pretreatment in
conventional AAS and ICP-OES respectively.