High-Performance Liquid Chromatographic Determination of TCNB(tetrachloronitrobenzene) in Potatoes R. J. Bushway, et al.; Journal of Liquid Chromatography & Related Technologies, Volume 7, Issue 6 May 1984 , pages 1185 - 1193
Abstract: A high-performance liquid chromatographic (HPLC) method was developed for the determination of TCNB (tetrachloronitrobenzene), a sprout inhibitor, in potato peels and flesh fortified at levels of 0.16 to 53.5 ppm. TCNB was analyzed on a u Bondapak C18 column with UV detection at 210 nm. The mobile phase was acetonitrile-methanol-water (35:35:30) at a flow rate of 1.0 ml/min. Retention time was approximately 10 min. TCNB was extracted by blending for 5 min in acetone. Samples at a level of 1 ppm or higher were directly injected whereas samples below 1 ppm were partitioned into hexane followed by passage through an alumina column. Average recoveries varied from 85.6 to 96.8% with coefficients of variation ranging from 2.18 to 11.68%. A study conducted to test 23 pesticides for possible interferences with TCNB demonstrated that none of them co-chromatographed. The lower limit of detection was 0.08 ppm.