B.
气相色谱(GC)
在GC分析方法描述中,至少要包括如下内容。如分析方法中用到了其它的一些参数,对此也应当要进行说明。如果分析方法开发表明只有某一商业来源的色谱柱是适用的,则在分析方法中应当要包括这些资料。如果有多种色谱柱都是适用的话,则应当要包括等效色谱柱列表。
1. Column
---Column dimensions: length, internal diameter, external diameter
---Stationary phase
---Column material (e.g., silica, glass, stainless steel)
--- Column conditioning procedure
2. Operating Parameters
---Gases: purity, flow rate, pressure
---Temperatures: column, injector, detector (including temperature program, if used)
---Injection (e.g., split, splitless, on-column)
---Detector
---Typical retention time and total run time
1.色谱柱
色谱柱尺寸:柱长,内外径
固定相
柱填充料(比如:硅酸,玻璃,不锈钢)
柱调节程序
2.操作参数
载气:纯度,流速,压力
温度:色谱柱,进样器,检测器(包括升温程序,如用到的话)
进样:(比如:分流,不分流,柱头进样(On-column injection)
检测限
典型保留时间和总运行时间。
3. System Suitability Testing
Appropriate system suitability criteria should be defined and included in all analytical procedures.
If an internal standard is used, the minimum acceptable resolution between the internal standard and one or more active ingredient should be specified. If the analytical procedure is used to control the level of impurities, the minimum resolution between the active ingredient and the closest eluting impurity, or the two peaks eluting closest to each other, should be given.
The RSD is normally performed at the beginning of the run. However, for assays with lengthy run times or as otherwise justified by the applicant, the reported average may be taken from injections at the beginning and end of the run, or beginning, middle, and end of the run.
3.系统适应性实验
在所有的分析方法中都应当要有适当的系统适应性要求。
如采用国内标准,则应满足内标和一个或多个组分的最小分离度要求。如果该分析方法是用于控制杂质水平的,则应当要说明活性成分和最邻近杂质组分,或每两相邻组分的分离度最小可接受值。
一般来说,在仪器开始运行时,进行进样精密度实验,计算其相对标准偏差。然而,对于运行时间很长的分析,或申请者另有理由说明,则可以在运行的开始和结束时,或在运行的开始,中间和结束时进样,然后报告其平均值。
C. Spectrophotometry, Spectroscopy, Spectrometry and Related Physical Methodologies
These analytical procedures include, but are not limited to, IR spectrophotometry, near IR spectrophotometry (NIR), UV/visible spectrophotometry (UV/Vis), atomic emission and atomic absorption, NMR, Raman spectroscopy, MS, and XRD.
Spectrometric analytical procedures may not be stability-indicating. The bias of the analytical procedure should be evaluated by comparing it with a chromatographic procedure, where appropriate. When manually operated equipment is used, the description of the analytical procedure should include an acceptance criterion for the amount of time that may elapse between sampling and reading. Appropriate system suitability and/or calibration testing is recommended. Validation criteria should include specificity (demonstrating no interference of placebo), linearity, repeatability, intermediate precision, and robustness.
C:分光光度法,光谱法和相关的物理方法
(IR),近红外光说(NIR),紫外可见光谱(UV/Vis),原子发射光谱/
原子吸收光谱,核磁共振(NMR),拉曼光谱,质谱(MS)和X-射线衍射(XRD)。
光谱分析方法可能没有稳定性指示能力。必要的话,可能通过与色谱方法的比较来评估光谱分析方法的偏差。如果用到了手动仪器的话,则分析方法描述中应当要有取样(sampling)和读数(reading)之间时间差的可接受标准。建议使用适当的系统适应性实验和/或校准实验。验证标准中应当要包括专属性(说明没有空白干扰),线性,重复性,中间精密度和耐用性。
D. Capillary Electrophoresis (CE)
At a minimum, the parameters listed below should be specified for a capillary electrophoretic analytical procedure. Additional parameters may be included as required by the procedure.
If method development has indicated that capillaries from only one commercial source are suitable, this information should be included as part of the analytical procedure. If more than one capillary is suitable, a listing of capillaries found to be equivalent should be included.
1. Capillary
--Capillary dimensions: length, length to detector, internal diameter, external
diameter
--Capillary material
--Capillary internal coating (if any)