主题:【讨论】ICP中的信背比如何做?

浏览0 回复13 电梯直达
依风1986
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 依风1986(xurunjiao5339) 发表:
2.5.1 Background Equivalent Concentration (BEC)
背景值
2.5.1.1 The BEC value is the concentration of an element, which would produce the
same emission intensity as the plasma background measured at the analyte
wavelength. The BEC serves as an indication of instrument sensitivity which is
related to torch alignment, plasma viewing height, nebulizer gas flow rate, and
incident RF power.
背景值是测量一浓度元素在分析波长下产生相应的发射强度。BEC可指示仪器的
灵敏度,它与炬管线性、等离子观测的高度,雾化气流量,射频功率有关。
2.5.1.2 The BEC should be determined every three months
BEC应每3个月测定一次。
2.5.1.3 Measure the intensity of a 5% nitric acid blank solution (IB).
测量5%的硝酸空白溶液的强度(IB)。
2.5.1.4 Measure the intensity of a 1.0 ppm Mn standard in 5% nitric acid (ISTD)at
a wavelength of 257.610.`
在波长为257.610nm处测量基体为5%硝酸的1.0ppm Mn标准溶液的强度。
2.5.1.5 Calculate the BEC = IB / (ISTD – IB) X CSTD
计算 BEC = IB / (ISTD – IB) X CSTD
Where: CSTD = concentration of Mn STD = 1.0 ppm
这里: CSTD=Mn 标准溶液浓度 = 1.0 ppm
IB and ISTD have been automatically corrected to the dark
current
IB和ISTD 已经被暗电流自动校正
Record the BEC value. The BEC must not be greater
than 0.04.
记录BEC的值。BEC不能大于0.04。
2.5.1.6 If the BEC is greater than 0.04, check the sample introduction system and
the torch for wear. Replace the torch and/or change o-rings and other worn
parts as needed. Reanalyze the BEC by following steps §2.5.1.1.3 to
§2.5.1.1.5. If the BEC still does not meet the acceptance limit, notify a
supervisor immediately and take the instrument out of service.
如果BEC大于0.04,检查进样系统和炬管装置。根据需要更换炬管和/或O形圈和
其他损耗部件。按照步骤§2.5.1.1.3至2.5.1.1.5重新分析BEC。如果BEC仍然
不符合接受限值,立即通知监管者和将仪器外修。
2.5.1.7 The BEC shall be charted to show any trending.
BEC应该标记显示出任何的变化趋势。
Notice: For PE ICP instrument, the dark current is obtained by closing the shutter.
注意: 对于PE的ICP,通过关闭快门获得暗电流。

这是MATTEL GLOP 0006-7409中对BEC的做法解释,可以参考下玩具老大的标准
依风1986
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
信背比还是按照常规去考虑,你的基质空白强度,你的溶液强度多少?
beyond1977
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 chemistryren(chemistryren) 发表:
我的仪器是PE2100,想做信背比,问了PE的人,说是配个1PPM的标准溶液,把它的强度和空白的强度比就是信背比..但为什么要拿1PPM做呢?如果用10PPM做出来信背比岂不是要高10倍??各位是怎么做这个信背比的呢?

信背比就应该这样做啊?我想你可能是想做最低检测限吧,同一个元素,一定范围内浓度高,信背比也会增高,但肯定不会完全成正比