请各位专家审阅你们的辅导成绩:在你们的精心辅导下,个人感觉水平有不小的进步,看看你们的学生有进步没?多提宝贵意见!
8.1.1.1 Alarm historic menu报警历史菜单
Displays the last 25 system alarms or peak value alarms.
显示最新的25个系统报警或者峰值报警。
System alarms :系统报警
􀂾 Low sample flow:样品气流量低报警
When the sample flow goes below 10 CC/min with a sample flow set point bigger than 10 CC/min;
当样品气流量低于10 CC/min时,就会发出流量低报警。这是因为样品流量设定值比10 CC/min大。…30秒后,系统会自动进行等待状态,也不再进样,以保护分析仪安全…
􀂾 Low carrier flow:载气流量低报警
When the carrier flow falls below 5 CC/min for the plasma and 2 CC/min for the TCD during 30 seconds, a “Plasma shut down” or “TCD shut down” alarm will be initiated after 30 seconds to protect the system by turning off the plasma or TCD.
对于等离子体来说,当载气流量降至5 CC/min;对TCD来说流量低于2 CC/min ,低流量在30秒内,30秒后将发出等离子体将关闭或者TCD关闭的报警信息,关闭等离子体或者热导检测器是为了保护系统。
􀂾 Plasma shut down or TCD shut down: 等离子体将关闭或者TCD关闭
When a “Low carrier flow” alarm remains active for 30 seconds, turns the plasma or TCD off.
当一个低载气流量报警出现有30秒时,等离子体或者TCD关闭。
􀂾 Plasma OFF:等离子体关闭
When the cell signal counts are lower than the starting count, it indicates that the plasma is physically off.
当检测器信号值比起动时数值低的时候,它代表着等离子体实际上已经自然关闭。
􀂾 Starting:启动
When a “Plasma off” alarm is active, no “Low carrier flow” alarm is active, the starting mode is automatic (see section 8.3.1 about the SYSTEM
CONFIGURATION MENU for the Starting mode definition) and no cycle is
progressing (real-time chromatogram stopped or between cycles), the plasma is restarted
当一个等离子体关闭的报警被激活的时候,但没有载气流量低的报警出现,等离子体会在无分析周期进行(实时分析或两个分析周期之间)自动重新启动。
详情请看系统配置菜单中关于起动模式的解释。
􀂾 Plasma ON等离子体打开
When the cell signal counts go above the starting count after a “Plasma OFF”, it indicates that the plasma is physically on.
当等 离子关闭以后的检测器池量已经超过启动池量的时候,它表示等离子体可以自然打开。
􀂾 RTD problem:温度控制器故障
When an oven temperature falls below 10 degrees Celsius.
当柱温听跌至10度以下的时候。
􀂾 Carrier flow deviation:载气流量偏差大
When the carrier flow is smaller or bigger by 2 CC/min from the carrier flow set point, the injection is still possible and the system continues to report process values.
当载气流量大于或小于设定值2 CC/min的时候,仪器依然进样,且系统继续报告处理结果。
􀂾 Oven temperature deviation:柱温偏差大
When an oven temperature is smaller or bigger by 1 degree Celsius from the oven temperature set point.
当柱温大于或小于设定值1度的时候。
该帖子作者被版主 何当奇8积分, 2经验,加分理由:有效求助
chengjingbao
Peak value alarms :峰值报警:
(no impact on the system status alarm dry contact output)
系统状态报警干式接点输出无效
􀂾 Alarm 1 peak # :峰值报警1
When the concentration of impurity is greater than the Alarm 1 for the peak.
当杂质含量比峰值报警1要大时。
􀂾 Alarm 2 peak # : 峰值报警2
When the concentration of impurity is greater than the Alarm 2 for the peak.
当杂质含量比峰值报警1要大时。
􀂾 Overscale peak # :峰值超量程范围
When the concentration of impurity is greater than the actual scale of the peak.
当杂质含量超过峰实际量程的时候。
When a problem arises, a new alarm is generated and the button of the ALARM HISTORIC MENU turns red. If the menu is opened and an alarm is still active, it turns yellow. If another alarm occurs, it turns red again. The button will turn green if all
alarms are resolved.
当一个问题出现的时候,产生一个新的报警,且报警历史菜单的按钮变为红的。如果打开菜单,报警依然存在。但它变为黄色。如果另一个报警出现了,它将又变为红色。如果全部的报警被解决,按钮将变绿。
Every system alarms will activate (or deactivate, depending the settings in the ALARM CONFIGURATION MENU) the system status alarm dry contact output. Peak value alarm 1 and 2 have each a separate dry contact output that is activated (or deactivated, still depending on the ALARM CONFIGURATION MENU) when the alarm
thresholds are reached.
每个系统报警通过系统状态报警干式接点输出激活(或取消,依靠报警配置菜单的设定)。当报警端到达的时候,有单独的系统状态报警干式接点输出的峰值报警1和2被激活(或取消,依靠报警配置菜单的设定)。
Some system functions such as a real-time chromatogram are disabled when alarms are active. Rectify the problems before proceeding with your analysis.
当报警被启动的时候,一些系统功能如实时分析被启动。在继续进行你的分析前,纠正你的问题。
When the problem is resolved, the same alarm message is displayed, but with an “Ok” message (except for a “Plasma shut down” that is resolved with a “Starting” and for the
“Plasma OFF” that is resolved with a “Plasma ON”).
当问题被解决的时候,同样的报警信息显示出来,仅是一个“OK信息”(除了“Plasma shut down”,解决方法是“Starting ”和“Plasma OFF”,解决方法是“Plasma ON”)。
Example:例如:
• If a “Low carrier flow” occurs, the following message is displayed:如果一个“载气流量低”出现,如下和信息将被显示
(date) Low carrier flow: (hour)
(数据)低载气流量。(小时)
• When the flow problem is rectified, the following message is displayed:当流量问题被调整的时候,如下信息将被显示。
(date) Low carrier flow: Ok (hour)
(数据)低载气流量:OK(小时)
何当奇
为什么不是“样品流量低”,多个“气”有什么作用?
样品流量设定值大于10 CC/min时, 如果样品流量低于10 CC/min,就..
后面那句没英文的?
何当奇
flow ,是不是就是flow rate? 流量翻译为流速吧。
貌似很多都不是完整的句子?
阿du
原文由 redanqi 发表:
后面那句没英文的?
红狐姐,老程测气体的,还是用样品气好一些,具体对应。
呵呵发了这么多,看了头晕了。
那个“流量低警报”改为流量过低警报更符合使用习惯和便于理解。
阿du
原文由 redanqi 发表:
貌似很多都不是完整的句子?
还是流量准确,流速一般指线速度的,通常我们使用的都是流量。
何当奇
原文由 duliuhui 发表:
原文由 redanqi 发表:
貌似很多都不是完整的句子?
哦,我这里翻译都是流速,可能行业不同吧.
其它的我中午有时间再看看
chengjingbao
原文由 redanqi 发表:
貌似很多都不是完整的句子?
是的,大部分是故障和报警判断。先讲现象,后讲原因,所以看上去都是半截的,上下一对照,就清楚了。
chengjingbao
原文由 duliuhui 发表:
原文由 redanqi 发表:
后面那句没英文的?
呵呵发了这么多,看了头晕了。
这等于是我的期中考试,让各位专家看看他们的辅导结果。流量过低报警可以,平时我也是这么叫的。因为有高报、低报之分,就写成低报警了。
happyjyl
原文由 duliuhui 发表:
是测气体样品还是做的是顶空气相
happyjyl
原文由 redanqi 发表:
原文由 duliuhui 发表:
原文由 redanqi 发表:
貌似很多都不是完整的句子?
其它的我中午有时间再看看
我平时接触到的中国药典、美国药典、欧洲药典和技术资料,里面一般都用flow rate,也没有特别指出是线速度。仅有一个资料中在提到载气流速时是用到了线速度:
The carrier gas is helium flowing with a linear velocity of about 11 cm per second.