原文由 yaomengnan 发表:
要怎么算检出线呢
1993年,IUPAC的规定:
The concentration or mass flow of a sample component in the mobile phase that gives a detector signal equal to twice the noise level. It can be calculated from the measured sensitivity (S) and noise (N):
D= 2N/S
where D is the minimum detectability, expressed either as concentration or mass-flow of the substance of interest in the mobile phase at the detector. Both sensitivity and minimum detectability must be determined for the same substance.