原文由 bewom 发表:
Argon is not flowing to the spectrometer.
Argon is needed by the spectrometer to protect the detector. The input pressure should be between 550 and 825 kPa (80 to 120 psi).
Check the pressure on the argon supply and the connections between the supply and the spectrometer. After the argon supply is restored, click on the Initialize button on the Spectrometer page of the Diagnostics window to begin warm-up. Warm up will require between 15 and 75 minutes depending on the length of time the spectrometer was disabled.
氩气没进入光谱仪
光谱仪需要氩气保护其检测器。输入压力应在550-850KPa.
检查氩气源压力和气瓶和光谱仪之间的连接。在恢复供气后,点击软件诊断窗口的光谱仪页面上的初始化按钮预热机器。根据光谱仪断气时间的长短,预热需要15-75分钟
PE的机器,原来碰到过几次,觉得PE的机子喜欢出点这个毛病。好像不是氩气源的问题,是空气压缩机压力不够或有个地方堵住了。