主题:【求助】Pb标准曲线

浏览0 回复30 电梯直达
xuemei86520
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007
为您推荐
您可能想找: 气相色谱仪(GC) 询底价
专属顾问快速对接
立即提交
可能感兴趣
ldgfive
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007

你的标准曲线浓度太不了,吸光度值应该也是非常小的,过小的吸光度肯定会出现波动
kjmmyj
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007



测完高标后要把仪器用稀酸冲洗一下,另外燃烧器可以拿下来用水冲洗一会,你的标准溶液吸光值也比较大
newxys
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
xuemei86520
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 ldgfive(ldgfive) 发表:
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007

你的标准曲线浓度太不了,吸光度值应该也是非常小的,过小的吸光度肯定会出现波动

因为我要测的样品浓度在10-100mg/kg之间,取样一般为0.1g-0.5g,浓度范围在上面标准溶液浓度范围之内,
chemistryren
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007

你是什么仪器啊??是用火焰测的吧??0.5PPM的吸光度有0.9之多??这么高的灵敏度精密度差也就不足为奇了.
yangyon8
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007


怎么会这样呢?浓度这么小,吸光值这么大。我的数据如下

也请各位参详。
qqgghh
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
xuemei86520
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 chemistryren(chemistryren) 发表:
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007

你是什么仪器啊??是用火焰测的吧??0.5PPM的吸光度有0.9之多??这么高的灵敏度精密度差也就不足为奇了.

是的,我用的是岛津6300c火焰原子吸收,那样怎样提高提高精密度,用这样的曲线测得的样品结果准确吗?
chemistryren
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 xuemei86520(xuemei86520) 发表:
原文由 chemistryren(chemistryren) 发表:
原文由 xuemei86520(xuemei86520) 发表:
我做铅标准曲线时遇到一些问题,想请教一下高手,我配的铅标准溶液是0,0.1ug/ml,0.5ug/ml,1ug/ml,做出来的标准曲线的相关系数是0.9999,但用这条曲线,测上面的标准溶液差别很大,结果为下:
0.1ug/ml:0.0871,0.0822,0.0625,0.1117,0.0822
0.5ug/ml:0.4021,0.4759,0.4513,0.4267,0.4415
1ug/ml  :0.9484,0.9238,0.9385,0.9533,0.8007

你是什么仪器啊??是用火焰测的吧??0.5PPM的吸光度有0.9之多??这么高的灵敏度精密度差也就不足为奇了.

是的,我用的是岛津6300c火焰原子吸收,那样怎样提高提高精密度,用这样的曲线测得的样品结果准确吗?

你的标准曲线浓度是不是搞错了??不可能有这么高的灵敏度的。
liangmaolin
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
建议富集样品中的pb,你的标样浓度太低,误差相对较大,我用的是1ppm. 2ppm  3ppm线性效果就相当的好。