主题:关于信噪比计算
浏览0 回复4 电梯直达
sonics
悬赏金额:10积分 状态: 已解决
各位老师好,本人使用岛津GCMS-TQ8040三重四级杆,在计算信噪比的时候,发现有几种信噪比的算法,ASTM,RMS和峰至峰,请哪位老师讲一讲,这几种信噪比计算的原理,以及什么情况下用哪一种信噪比的计算方法,为什么???
非常感谢!
推荐答案:安平回复于2015/12/04
Noise Calculation Methods
There are three methods for calculating noise:ASTM, RMS, and Peak to Peak.
ASTM
A 15-minute baseline period is first separated into 30 second sections. Noise width of each section is individually calculated as described below. Then the noise widths of all sections are averaged to obtain the ASTM noise.
[Calculation procedure]
(1) The slope of the approximated linear line for all data points within each section is calculated by the least square method.
(2) Two parallel lines are assumed by shifting the slope upward and downward, so that all data points within the section fall between two lines.
(3) The distance between the two parallel lines in the intensity axis direction is defined as the noise width for the section.
RMS
The RMS noise refers to the standard deviation of the measurement point signals detected within a specified section.
Peak to Peak
"Noise" refers to the value of the "maximum intensity value - minimum intensity value" in the noise time range.
补充答案:
千层峰回复于2015/12/02
岛津GCMS-TQ8040三重四级杆没用过,,你看看下边链接中的帖子。。
http://www.instrument.com.cn/search/search_bbs?keywords=%E4%BF%A1%E5%99%AA%E6%AF%94&loginSource=15&tabType=2&position=&forumid=44&PaperType=&search_sort=&FromDate=&ToDate=
该帖子作者被版主 symmacros2积分, 2经验,加分理由:发帖
千层峰
安平
ASTM
RMS
Peak to Peak
期盼奇迹的孩子
安捷伦的算是用哪种方法,每年仪器校验时都是自动算的,还没怎么注意过方法~~~
Ins_71a3db04
原文由 安平(byron1111) 发表:Noise Calculation MethodsThere are three methods for calculating noise:ASTM, RMS, and Peak to Peak.ASTMA 15-minute baseline period is first separated into 30 second sections. Noise width of each section is individually calculated as described below. Then the noise widths of all sections are averaged to obtain the ASTM noise.[Calculation procedure](1) The slope of the approximated linear line for all data points within each section is calculated by the least square method.(2) Two parallel lines are assumed by shifting the slope upward and downward, so that all data points within the section fall between two lines.(3) The distance between the two parallel lines in the intensity axis direction is defined as the noise width for the section.RMSThe RMS noise refers to the standard deviation of the measurement point signals detected within a specified section.Peak to Peak"Noise" refers to the value of the "maximum intensity value - minimum intensity value" in the noise time range.
老师,请问我们计算噪声的时候要选择哪种计算方式呀,这三种计算方式算出来的S/N差别还挺大的
手机版: 关于信噪比计算