Tools needed (需要工具)
- allen keys  (内六角扳手)
- Voltmeter   (电压表)

Procedure (步骤)
Step 1: Switch OFF the X-Ray tube.    (关光管)
Step 2: Open the right back panel (for Advant’XP) and the right one (for 9800XP) 
        【打开右后面板 (Advant’XP系列) 或者打开右面板(9800XP系列)  】
Step 3: On the figure , you will find the 2 floats. (上图右两个浮漂)
为您推荐
专属顾问快速对接
立即提交
Onway
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
For primary water sensor adjustment:

Step 4.1: Plug the voltmeter on TS1 connector (pin 3 & 4), you should get ~5VDC
Step 5.1: Enter the command PW 1 in order to switch on the primary water flow. The floatshould go up and the voltage on TS1 go to ~0VDC
Step 6.1: unscrew a little bit the 2 imbus screws and goes up until you get 5VDC. Put a Mark and go down. You will again get 0VDC and still go down you will again get 5VDC. Again put another mark. The position to adjust the sensor is exactly in the middle position of the 2 mark.

For secondary water sensor adjustment:

Step 4.2: Plug the voltmeter on TS1 connector (pin 1 & 2), you should get ~5VDC
Step 5.2: Enter the command SW 1 in order to switch on the secondary water flow. The float should go up and the voltage on TS1 go to ~0VDC
Step 6.2: unscrew a little bit the 2 imbus screws and goes up until you get 5VDC. Put a Mark and go down. You will again get 0VDC and still go down you will again get 5VDC. Again put another mark. The position to adjust the sensor is exactly in the middle position of the 2 mark.
Step 7: install back the panel and switch on the X-Ray tube