For primary water sensor adjustment:
Step 4.1: Plug the voltmeter on TS1 connector (pin 3 & 4), you should get ~5VDC
Step 5.1: Enter the command PW 1 in order to switch on the primary water flow. The floatshould go up and the voltage on TS1 go to ~0VDC
Step 6.1: unscrew a little bit the 2 imbus screws and goes up until you get 5VDC. Put a Mark and go down. You will again get 0VDC and still go down you will again get 5VDC. Again put another mark. The position to adjust the sensor is exactly in the middle position of the 2 mark.
For secondary water sensor adjustment:
Step 4.2: Plug the voltmeter on TS1 connector (pin 1 & 2), you should get ~5VDC
Step 5.2: Enter the command SW 1 in order to switch on the secondary water flow. The float should go up and the voltage on TS1 go to ~0VDC
Step 6.2: unscrew a little bit the 2 imbus screws and goes up until you get 5VDC. Put a Mark and go down. You will again get 0VDC and still go down you will again get 5VDC. Again put another mark. The position to adjust the sensor is exactly in the middle position of the 2 mark.
Step 7: install back the panel and switch on the X-Ray tube