The basic principles for validation and certification of a sterilizing process are enumerated as follows:
1. Establish that the process equipment has capability of operating within the required parameters.
2. Demonstrate that the critical control equipment and instrumentation are capable of operating within the prescribed parameters for the process equipment.
3. Perform replicate cycles representing the required operational range of the equipment and employing actual or simulated product. Demonstrate that the processes have been carried out within the prescribed protocol limits and finally that the probability of microbial survival in the replicate processes completed is not greater than the prescribed limits.
4. Monitor the validated process during routine operation. Periodically as needed, requalify and recertify the equipment.
5. Complete the protocols, and document steps (1) through (4) above.
The principles and implementation of a program to validate an aseptic processing procedure are similar to the validation of a sterilization process. In aseptic processing, the components of the final dosage form are sterilized separately and the finished article is assembled in an aseptic manner.
Proper validation of the sterilization process or the aseptic process requires a high level of knowledge of the field of sterilization and clean room technology. In order to comply with currently acceptable and achievable limits in sterilization parameters, it is necessary to employ appropriate instrumentation and equipment to control the critical parameters such as temperature and time, humidity, and sterilizing gas concentration, or absorbed radiation. An important aspect of the validation program in many sterilization procedures involves the employment of biological indicators (see Biological Indicators for Sterilization 1035 ). The validated and certified process should be revalidated periodically; however, the revalidation program need not necessarily be as extensive as the original program.
A typical validation program, as outlined below, is one designed for the steam autoclave, but the principles are applicable to the other sterilization procedures discussed in this informational chapter. The program comprises several stages.
The installation qualification stage is intended to establish that controls and other instrumentation are properly designed and calibrated. Documentation should be on file demonstrating the quality of the required utilities such as steam, water, and air. The operational qualification stage is intended to confirm that the empty chamber functions within the parameters of temperature at all of the key chamber locations prescribed in the protocol. It is usually appropriate to develop heat profile records, i.e., simultaneous temperatures in the chamber employing multiple temperature-sensing devices. A typical acceptable range of temperature in the empty chamber is ±1 when the chamber temperature is not less than 121 . The confirmatory stage of the validation program is the actual sterilization of materials or articles. This determination requires the employment of temperature-sensing devices inserted into samples of the articles, as well as either samples of the articles to which appropriate concentrations of suitable test microorganisms have been added, or separate BIs in operationally fully loaded autoclave configurations. The effectiveness of heat delivery or penetration into the actual articles and the time of the exposure are the two main factors that determine the lethality of the sterilization process. The final stage of the validation program requires the documentation of the supporting data developed in executing the program.
It is generally accepted that terminally sterilized injectable articles or critical devices purporting to be sterile, when processed in the autoclave, attain a 10–6 microbial survivor probability, i.e., assurance of less than 1 chance in 1 million that viable microorganisms are present in the sterilized article or dosage form. With heat-stable articles, the approach often is to considerably exceed the critical time necessary to achieve the 10–6 microbial survivor probability (overkill). However, with an article where extensive heat exposure may have a damaging effect, it may not be feasible to employ this overkill approach. In this latter instance, the development of the sterilization cycle depends heavily on knowledge of the microbial burden of the product, based on examination, over a suitable time period, of a substantial number of lots of the presterilized product.
The D value is the time (in minutes) required to reduce the microbial population by 90% or 1 log cycle (i.e., to a surviving fraction of 1/10), at a specific temperature. Therefore, where the D value of a BI preparation of, for example, Bacillus stearothermophilus spores is 1.5 minutes under the total process parameters, e.g., at 121 , if it is treated for 12 minutes under the same conditions, it can be stated that the lethality input is 8D. The effect of applying this input to the product would depend on the initial microbial burden. Assuming that its resistance to sterilization is equivalent to that of the BI, if the microbial burden of the product in question is 102 microorganisms, a lethality input of 2D yields a microbial burden of 1 (10 theoretical), and a further 6D yields a calculated microbial survivor probability of 10–6. (Under the same conditions, a lethality input of 12D may be used in a typical “overkill” approach.) Generally, the survivor probability achieved for the article under the validated sterilization cycle is not completely correlated with what may occur with the BI. For valid use, therefore, it is essential that the resistance of the BI be greater than that of the natural microbial burden of the article sterilized. It is then appropriate to make a worst-case assumption and treat the microbial burden as though its heat resistance were equivalent to that of the BI, although it is not likely that the most resistant of a typical microbial burden isolates will demonstrate a heat resistance of the magnitude shown by this species, frequently employed as a BI for steam sterilization. In the above example, a 12-minute cycle is considered adequate for sterilization if the product had a microbial burden of 102 microorganisms. However, if the indicator originally had 106 microorganisms content, actually a 10–2 probability of survival could be expected; i.e., 1 in 100 BIs may yield positive results. This type of situation may be avoided by selection of the appropriate BI. Alternatively, high content indicators may be used on the basis of a predetermined acceptable count reduction.