934的:
Limit of benzene— Proceed as directed in the test for Limit of benzene under
Carbomer 934P, except to use about 20 mg of Carbomer 934, accurately weighed, instead of about 1 g of Carbomer 934P, to prepare the Test solution. Calculate the percentage of benzene in the portion of Carbomer 934 taken by the formula:
10(C/W)(rU / rS)
in which C is the concentration, in µg per mL, of benzene in the Standard solution; W is the weight, in mg, of Carbomer 934 taken to prepare the Test solution; and rU and rS are the benzene peak responses obtained from the Test solution and the Standard solution, respectively: not more than 0.5% is found.
934P的:
Standard solution— Quantitatively dissolve an accurately weighed quantity of benzene in methanol to obtain a solution having a known concentration of about 0.2 mg per mL. Quantitatively dilute this solution with organic-free water (see Residual Solvents [img=12,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/lang.gif[/img]467[img=12,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/rang.gif[/img]) to obtain a solution having a known concentration of about 1.0 µg per mL.
Test solution— Transfer about 1 g of Carbomer 934P, accurately weighed, to a 100-mL volumetric flask. Add about 75 mL of sodium chloride solution (2 in 100), and mix by mechanical means until homogenous (usually about 30 minutes). Dilute with sodium chloride solution (2 in 100) to volume, and mix until homogenous (usually less than 1 minute). [note—This preparation must be analyzed within 3 hours of preparation.]
Chromatographic system (see
Chromatography [img=12,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/lang.gif[/img]621[img=12,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/rang.gif[/img])— The gas chromatograph is equipped with a flame-ionization detector, a 0.53-mm × 30-m fused silica analytical column coated with a 3.0-µm G43 stationary phase, a 0.53-mm × 5-m silica guard column deactivated with phenylmethyl siloxane, and a splitless injection system. The carrier gas is helium flowing at a linear velocity of about 35 cm per second. The injection port and detector temperatures are maintained at 140[img=7,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/deg.gif[/img] and 260[img=7,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/deg.gif[/img], respectively. The column temperature is programmed according to the following steps: it is held at 40[img=7,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/deg.gif[/img] for 20 minutes, then increased rapidly to 240[img=7,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/deg.gif[/img], and maintained at 240[img=7,20]mk:@MSITStore:E:\药典\USP\usp32_nf27_2009_roydan.chm::/pub/data/images/chars/deg.gif[/img] for 20 minutes. Chromatograph the Standard solution, and record the responses as directed for Procedure: the relative standard deviation for replicate injections is not more than 15%.
Procedure— Separately inject equal volumes (about 1 µL) of the Standard solution and the Test solution into the chromatograph, record the chromatograms, and measure the responses for the benzene peaks. Calculate the percentage of benzene in the portion of Carbomer 934P taken by the formula:
10(C/W)(rU / rS)
in which C is the concentration, in µg per mL, of benzene in the Standard solution; W is the weight, in mg, of Carbomer 934P taken to prepare the Test solution; and rU and rS are the benzene peak responses obtained from the Test solution and the Standard solution, respectively: not more than 0.01% is found.
940:
Carbomer 934P, except to use about 20 mg of Carbomer 940, accurately weighed, instead of about 1 g of Carbomer 934P, to prepare the Test solution. Calculate the percentage of benzene in the portion of Carbomer 940 taken by the formula:
10(C/W)(rU / rS)
in which C is the concentration, in µg per mL, of benzene in the Standard solution; W is the weight, in mg, of Carbomer 940 taken to prepare the Test solution; and rU and rS are the benzene peak responses obtained from the Test solution and the Standard solution, respectively: not more than 0.5% is found.
941:
Carbomer 934P, except to use about 20 mg of Carbomer 941, accurately weighed, instead of about 1 g of Carbomer 934P, to prepare the Test solution. Calculate the percentage of benzene in the portion of Carbomer 941 taken by the formula:
10(C/W)(rU / rS)
in which C is the concentration, in µg per mL, of benzene in the Standard solution; W is the weight, in mg, of Carbomer 941 taken to prepare the Test solution; and rU and rS are the benzene peak responses obtained from the Test solution and the Standard solution, respectively: not more than 0.5% is found.
1342:
Carbomer 934P, except to use about 50 mg of Carbomer 1342, accurately weighed, instead of about 1 g of Carbomer 934P, to prepare the Test solution. Calculate the percentage of benzene in the portion of Carbomer 1342 taken by the formula:
10(C/W)(rU / rS)
in which C is the concentration, in µg per mL, of benzene in the Standard solution; W is the weight, in mg, of Carbomer 1342 taken to prepare the Test solution; and rU and rS are the benzene peak responses obtained from the Test solution and the Standard solution, respectively: not more than 0.2% is found.
好象因为限度不同只是称样量上有所不同,没有很大的差别