原文由 依风1986(xurunjiao5339) 发表:
2.5.1 Background Equivalent Concentration (BEC)
背景值
2.5.1.1 The BEC value is the concentration of an element, which would produce the
same emission intensity as the plasma background measured at the analyte
wavelength. The BEC serves as an indication of instrument sensitivity which is
related to torch alignment, plasma viewing height, nebulizer gas flow rate, and
incident RF power.
背景值是测量一浓度元素在分析波长下产生相应的发射强度。BEC可指示仪器的
灵敏度,它与炬管线性、等离子观测的高度,雾化气流量,射频功率有关。
2.5.1.2 The BEC should be determined every three months
BEC应每3个月测定一次。
2.5.1.3 Measure the intensity of a 5% nitric acid blank solution (IB).
测量5%的硝酸空白溶液的强度(IB)。
2.5.1.4 Measure the intensity of a 1.0 ppm Mn standard in 5% nitric acid (ISTD)at
a wavelength of 257.610.`
在波长为257.610nm处测量基体为5%硝酸的1.0ppm Mn标准溶液的强度。
2.5.1.5 Calculate the BEC = IB / (ISTD – IB) X CSTD
计算 BEC = IB / (ISTD – IB) X CSTD
Where: CSTD = concentration of Mn STD = 1.0 ppm
这里: CSTD=Mn 标准溶液浓度 = 1.0 ppm
IB and ISTD have been automatically corrected to the dark
current
IB和ISTD 已经被暗电流自动校正
Record the BEC value. The BEC must not be greater
than 0.04.
记录BEC的值。BEC不能大于0.04。
做MATTEL的方法我就按照7409的要求去做
应该是concentration
原文由 tang566(tang566) 发表:原文由 依风1986(xurunjiao5339) 发表:
背景值
incident RF power.
BEC应每3个月测定一次。
测量5%的硝酸空白溶液的强度(IB)。
current
IB和ISTD 已经被暗电流自动校正
than 0.04.
应该是concentration
IB和ISTD 已经被暗电流自动校正,
那IB / (ISTD – IB)的值怎么计算?