主题:【讨论】有对AB的时间延迟碎裂了解的比较透彻的么
浏览0 回复7 电梯直达
kkeats
如题,AB的Qtrap特有的时间延迟碎裂(TDF)资料比较少,不知道怎么回事,论坛里不太冒头的高人请现身。。。。。。
该帖子作者被版主 sukiliang2积分, 2经验,加分理由:鼓励发帖
人散曲终
Time Delayed Fragmentation(TDF) working process:
1. Precursor ion selected in Q1
2. Ions are trapped in Q2(no fragmentation)
3. Ions are accelerated into the trap and fragment through collisionally activated dissociation
4. During the "cool"time,fragments below a user-specified mass are not trapped
5. After the "cool"time,all ions are collectd and scanned out of the trap
6. As ions are being scanned out of the trap,incoming ions are stored in the Q0 region(Q0 trapping)
I wish it would help you a lot.:)
该帖子作者被版主 sukiliang2积分, 2经验,加分理由:应助
sukiliang
kkeats
原文由 人散曲终(adam1943) 发表:
第三步 离子怎么碎的?碰撞气哪里来的?是在Trap的?尤其第四步,我记得一些资料上说,cool time容易再次碎裂离子排出离子阱,留下仅是母离子真正子离子,而非离子,不懂,求指点
人散曲终
原文由 kkeats(kkeats) 发表:
The TDF Mode is similar to Enhance Product Ion Scan,which perform MS/MS scan and Q3 becomes a filter.Fragment the ions in trap by setting Entry barrie between Q2&Q3 (TDF CE...Analyst advanced MS setting),no frag in Q2,Q2 is only used for trap,Q3 is used for frag/trap/scan.
When you set the Q3 Fill Mass in Analyst,during the "cool"time,fragments below the setting value are not trapped.After the "cool"time,all ions are collectd and scanned out of the trap
The grand-daughter ions is produced by MS3 mode.
该帖子作者被版主 nankingee5积分, 2经验,加分理由:讲解很清晰!
2013/4/6 0:46:54 Last edit by adam1943
netsking
这有两篇比较早的文献, 可供参考.
不知道如何upload PDF, 只好让大家去下吧.
http://medchem.rutgers.edu/AnalMedChem511/pdf_files/RB_pdf/LIT%20and%20QQQ.pdf
http://onlinelibrary.wiley.com/doi/10.1002/rcm.1061/pdf
该帖子作者被版主 nankingee5积分, 2经验,加分理由:感谢分享
netsking
记得在Training时, 说TDF主要用在做多肽,也就没有在细问了.
kkeats