主题:【讨论】为什么同样的方法和样品5973比5975的峰高和峰面积要大???

浏览0 回复64 电梯直达
可能感兴趣
jrp26818
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
我是用5973的。请教大侠,怎样才能延长MS中信号检测器(电子打拿极)使用寿命?
天马行空
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
wjl0220
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
linda7872
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
原文由 weifengtwo 发表:
After cleaning the source, I suddenly have no baseline on the chromatogram. Why?
If the problem occurred after the source was cleaned, it could be as a result of the high multiplier voltage caused by the dirty source, which could cause an excessive signal level when the Mass Spec is initially tuned or scanned after the clean source is installed.
To correct this problem, lower the Multiplier voltage to approximately 1000 volts in [MAUAL TUNE] and save this value to, “[TUNE PARAMETERS]”, in the ATUNE.U tune file and re-tune the Mass Spec. If the Mass Spec passes the tune, more than likely the MS is okay.

洗(离子)源后,发现基线突然消失了。不知为何?
若源清洗干净后发生这样的问题,可能是未清洁的源造成了EM高电压,而重新安装了清洁后的源时,MS初始调协电压偏高。要纠正这个问题,可在手动模式下将EM电压降低约1000V,并将该设置值保存到ATUNE.U调协文件的“调协参数”中,并重新调协,如调协能通过,问题应该就解决了。
怪侠一把刀
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
继续把这个帖子做下去!
希望能做成一个全面的基础问题和维护帖!望XDJM们相互拾柴抬庄啊!
How do I vent the MSD when my Personal Computer is not available?
Q:当计算机与仪器连接故障时,如何关闭质谱?

Answer:
If your MSD has a Gauge controller, turn it off. From the gc Keyboard turn OFF the gc Zone controlling the gc/MSD interface heater, for example AUX Temp. Also, turn the Inlet Temperature OFF and set the Oven Temperature to 30 degrees. Then disconnect the HIVAC Power Cable from the rear of the MSD. This turns off the MSD heater and Diffusion Pump heater or Turbo Pump, but the cooling fans and fore line pumps stay on. After 45 Minutes turn OFF the MSD Power Switch. Please refer to the 5973A Hardware Manual for further reference.
yanzi_0226
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
我同意楼上是说法!
可我现在做十溴联苯醚遇到了麻烦
因为把离子源温度设为230
可打进标样后检测不到十溴联苯醚
很着急啊
要是看到我的贴子能不能尽快给我回复啊
非常感谢啊!
yanzi_0226
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
下列是可能的原因为被观察的症状。 请检查。  -扩散泵是否或涡轮泵浦停止了或不满足标准操作条件?* -温度传感器没有连接?* -温度传感器不工作?* -来源是否是电源缆绳没有连接到弦面板?* - MSD电子单位正确地运作? *这些情况造成一个错误信息出现。 1. 检查真空状态。 > 看分开的项目。 2. 设法再联接缆绳为温度加热器传感器。  对绝缘的缆绳绿色小珠。 对白色织品被剥皮的缆绳,如果没有改善,替换温度传感器或与Axx顾客服务联系。
你看翻译得怎么样啊
我拿词典翻了一晚上哦!
yanzi_0226
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
The following are possible causes for the observed symptoms. Please check.
- Has the diffusion pump or turbo pump stopped or does not satisfy the standard operating conditions?*
- Is the temperature sensor is not connected?*
- Is the temperature sensor is not working?*
- Is the source power supply cable is not connected to the side board?*
- Is the MSD electronics unit working correctly?
* These situations cause an error message to appear.
1. Check the vacuum state. > See separate item. 2. Try re-connecting the cable for the temperature heater sensor.
Cable insulated by pairs of green beads. Pair of white fabric-skinned cables
If there is no improvement, replace the temperature sensor or contact Axx customer service.
下列是可能的原因为被观察的症状。 请检查。  -扩散泵是否或涡轮泵浦停止了或不满足标准操作条件?* -温度传感器没有连接?* -温度传感器不工作?* -来源是否是电源缆绳没有连接到弦面板?* - MSD电子单位正确地运作? *这些情况造成一个错误信息出现。 1. 检查真空状态。 > 看分开的项目。 2. 设法再联接缆绳为温度加热器传感器。  对绝缘的缆绳绿色小珠。 对白色织品被剥皮的缆绳,如果没有改善,替换温度传感器或与Axx顾客服务联系。
怪侠一把刀
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
谢谢楼上的辛苦劳作!
在您的基础上在用实际的语言转换了一下,方便理解!
再次感谢!给我继续的动力
The temperature of the ion source or quadrupole is unstable and shows 512 degrees. The message 'temperature control abnormal' appears. What could be the problem?
解决方法如下:
• The following are possible causes for the observed symptoms. Please check.
- Has the diffusion pump or turbo pump stopped or does not satisfy the standard operating conditions?*
- Is the temperature sensor is not connected?*
- Is the temperature sensor is not working?*
- Is the source power supply cable is not connected to the side board?*
- Is the MSD electronics unit working correctly?
* These situations cause an error message to appear.
1. Check the vacuum state. > See separate item. 2. Try re-connecting the cable for the temperature heater sensor.
Cable insulated by pairs of green beads. Pair of white fabric-skinned cables
If there is no improvement, replace the temperature sensor or contact Axx customer service.

问题:离子源或者四级杆温度显示故障情况?
答案:
下列是可能的原因为被观察的症状。 请检查。
-扩散泵是否或涡轮泵停止了或不满足标准操作条件?
-温度传感器没有连接?
* -温度传感器不工作?
* -离子源的侧板电源是否接好?
* - MSD电板是否烧坏OR能否正确地运作?
*这些情况造成一个错误信息出现。 1. 检查真空状态。 > 检查各独立的部件。 2. 离子源的绿、白色接线要检查是否接错,如果没有改善,替换温度传感器或与Axx顾客服务联系。

怪侠一把刀
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
7月25号  新问题
•Why does the library search results report the same compound more than once?

解答
The NIST, WILEY, and other libraries were not produced by Agilent. They contain spectra produced using magnetic field type equipment and by MS from various different manufacturers. As the library may contain a number of different measurements for the same compound, the compound may be displayed more than once in the results.
In the extended data analysis screen, open the [Spectra] menu and click [Edit Search Strategy] to open the [Search Strategy] dialog box. Selecting the [Delete duplications and CAS number] checkbox will prevent the same compound from appearing more than once.