主题:【原创】离子色谱忘记关机

浏览0 回复2 电梯直达
yfytap
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
戴安的1500外接rfc30,周五样多一直开着,后来忘了关机。周一去实验室发现主机和rfc都已关闭只是还通着电。后来正常开机发现电导很高15000多,检查了其他部分,没有发现异常,接着以正常流速压力浓度运行,经过连续2天20多个小时的运行,绝对电导降到了1.8左右,今天继续运行。

总结这次的问题:

程序文件:Pressure.LowerLimit =  200 [psi]
Pressure.UpperLimit =  3000 [psi]
%A.Equate =  "20mM KOH"
Pump_InjectValve.State  LoadPosition
Pump_ECD.Data_Collection_Rate =  5.0 [Hz]
CellTemperature.Nominal =  35.0 [°C]
ColumnTemperature.Nominal =  30.0 [°C]
Suppressor_Type =  ASRS_4mm
; Pump_ECD.Carbonate =  0.0
; Pump_ECD.Bicarbonate =  0.0
; Pump_ECD.Hydroxide =  20.0
; Pump_ECD.Tetraborate =  0.0
; Pump_ECD.Other eluent =  0.0
; Pump_ECD.Recommended Current =  50
Suppressor_Current =  50 [mA]
Channel_Pressure.Step =  0.20
Channel_Pressure.Average =  Off
Flow =  1.00 [ml/min]

-2.300 Pump_ECD_Relay_1.Closed Duration=138.00

0.000 Pump_ECD.Autozero
ECD_1.AcqOn
Channel_Pressure.AcqOn
Pump_InjectValve.InjectPosition Duration=30.00

11.40 ECD_1.AcqOff
Channel_Pressure.AcqOff

End
后来发现高纯水在130ml的时候主机已经关闭。想来是压力超下限了。而rfc30后来经咨询厂家工程师,模式在ttl control下自动关机对rfc30也没有影响。

所以这次的意外似乎影响不大。
为您推荐
您可能想找: 离子色谱(IC) 询底价
专属顾问快速对接
立即提交
lczhfj
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
有时意外情况出现让人手足无措,但未必都会造成大损伤,因有些仪器有保护功能。但细心仍是需要的,尽力避免不规范操作。此可供大家参考。
seafling
结帖率:
100%
关注:0 |粉丝:0
新手级: 新兵
应该没什么影响,流动相流感的时候,管子里是空气,淋洗液发生器即使开着也不会起作用,仪器无法构成电解回路